jc101:
It seems there is a recall for the QPX750SP Power Supply Serial Numbers 580064 to 585809.

A link to the recall pdf is here -> https://resources.aimtti.com/docs/Aim-TTi_QPX750SP_Recall_Notice_Mar2024.pdf

Text of the recall notice


--- Quote ---07/03/2023
Subject: Urgent Product Recall Notice - QPX750SP Mains Transformers

Dear Valued Customers,

We regret to inform you that a recent quality assessment has revealed an issue with the mains transformers used in the QPX750SP DC Laboratory Power Supply. After thorough investigation and disassembly of samples, it has come to our attention that the transformer supplier did not adhere to the precise specifications outlined in the engineering drawings.

Issue Details:
• Affected Product: QPX750SP
• Component: Mains Transformer
• Problem: Non-compliance with safety requirements due to incorrect build
• Risk Assessment: Although the transformers shipped have successfully passed the required product safety tests including the 2KV isolation test, the construction does not meet the defined safety standards.

Immediate Action Required: We take safety seriously, and in light of this discovery, we are initiating an urgent product recall for the following units.

Sno 580064 to 585809

Purchased between March 2023 and February 2024

Next Steps:

1. Stop Usage: If you have any QPX750SP units with the specified serial numbers, please immediately cease their usage.
2. Contact Us: Reach out to our customer support team at: Email sales@aimtti.com or Phone +44 (0)1480 412451 to arrange for the return and replacement of the affected unit.
3. Replacement Process: We will expedite the replacement process to minimize any disruption to your operations.

We sincerely apologize for any inconvenience this may cause and appreciate your cooperation in ensuring the safety of operation. Our commitment to quality remains unwavering, and we are committed to rectifying this situation promptly.

Thank you for your understanding.
--- End quote ---

woody:
My QPX7540 was affected by this problem and has been updated since.

The process was painless. I contacted AIM-TTI. Pick-up by UPS, return a couple of weeks later.

The only snag was that UPS missed that this was a return on a repair so they charged me some customs fee. We have Brexit to thank for that I suppose.

Anyways, the supply is back and working again. And supposedly safe :-)


 