Products > Test Equipment
AN8008 US $19, 9999count, 1uV, 0.01uA, 0.01Ohm, 1pF resolution meter
Mark Hennessy:
--- Quote from: ocw on July 16, 2017, 07:31:17 pm ---
--- Quote ---That isn't the "limitations of both meters influencing the other". Rather, that's the extra resistance of the measurement meter slightly reducing the current in the circuit to the DUT.
--- End quote ---
Let's take a specific example. I took my voltage reference which has a 5.00377 volt output and connected my 100 ohm 0.005% resistor (Y1453100R000V9L) in series with it. On my Agilent 34401A meter I measured 47.4017 mA. Under identical circumstances my AN8008 measured 49.9mA! That's because the shunt resistance on the 34401A is 5.022 ohms as compared to 0.0301 ohms on the AN8008. If you put both meters in series I would expect both meters to read about 47.4 mA due to the 34401A's shunt resistance. Which method is more meaningful and/or accurate? I want to know how accurate the current measurement is compared to the current without the meter inserted (50.0377 mA).
In the above situation the AN8008 was more accurate than my relatively expensive 34401A. However, if you correct the 34401A's reading, its accuracy is -0.016% as compared to the AN8008's accuracy of
-0.2452%. I'm trying to find the meter which has the best accuracy without requiring me to use my "slide rule."
--- End quote ---
Sorry, I feel like I'm clearly missing something really obvious.
Why does the actual current matter? Why go to the effort of using a precise voltage and a precise resistance in the hope of achieve a particular current when the total resistance added by the meter will be affected by the fuse (which is non-linear) and connectors/wires in addition to the shunt?
Or to put it another way, what is wrong with what I'm doing?
As I say, I don't care about the actual value of current - I just want to see how it compares to my known-good meters at certain approximate points, and to achieve that goal, I can't see any flaws with putting two or more meters in series. No slide rule required.
A current meter is specified in terms of the current passed through it, not the reading that results when some arbitrary voltage source is connected to it via some arbitrary resistance, no matter how precise those might be.
Obviously any current meter will have a shunt (and fuse, etc), so the current will be affected when you insert it into a circuit. The value of this resistance is just a fact of life, and is hardly a "figure of merit" given that all current meters must have a shunt to work. Just about any measurement will change things to a small extent; it's up to the operator to consider the effect of the particular shunt given the context: often it simply doesn't matter; other times, the burden voltage might indeed be a problem. But all that is quite separate from determining the basic accuracy of a current meter.
Fungus:
--- Quote from: boffin on July 16, 2017, 05:50:36 pm ---The AN8008 looks great except the two missing mA ranges. Does anyone know if the AN8002 suffers from the same fault?
--- End quote ---
No, it suffers from completely different ones.
Fungus:
--- Quote from: stj on July 16, 2017, 06:09:36 pm ---interesting thing in the manual for people who bitch about the fuses,
the maximum voltage for current readings is 36v DC or 25v AC.
--- End quote ---
You mean the maximum 10 amps will pass at 36V?
That's how Ohm's law works, yes, and you'll find similar values for other meters, even Flukes!
The usual workaround is to put something else in series with the multimeter. That way the voltage drop across the multimeter will only be a tiny fraction of the total voltage from the supply and (hopefully) much less than 36V. ;)
ocw:
--- Quote ---Or to put it another way, what is wrong with what I'm doing?
--- End quote ---
I thought that my example made it obvious. Doing it your way both of my meters connected in series read about 47.4 mA. Doing it my way one meter reads 49.9 mA while the other reads 47.4 mA. It's clear which one is more accurate in measuring current when you do not want to go to the trouble of correcting for the meter's shunt resistance.
Let's say that my 5.00377 is a critical voltage supply which can't dip below 5.0 volts. Connecting the 34401A will lower the voltage to about 4.7658 volts while the AN8008 insertion will only cause the voltage to drop to 5.00227 volts.
While my expensive electrometers can measure current with insignificant insertion resistance (or burden voltage), my relatively inexpensive ADA4530-1 current to voltage converter can do the same thing. Because of that I want the manufacturers to start adding that feature in more moderately priced meters. If you shrug your shoulders and say that you are happy the way things are nothing will change.
The Soulman:
--- Quote from: Fungus on July 16, 2017, 08:24:44 pm ---
--- Quote from: stj on July 16, 2017, 06:09:36 pm ---interesting thing in the manual for people who bitch about the fuses,
the maximum voltage for current readings is 36v DC or 25v AC.
--- End quote ---
You mean the maximum 10 amps will pass at 36V?
That's how Ohm's law works, yes, and you'll find similar values for other meters, even Flukes!
The usual workaround is to put something else in series with the multimeter. That way the voltage drop across the multimeter will only be a tiny fraction of the total voltage from the supply and (hopefully) much less than 36V. ;)
--- End quote ---
It has everything to do with the voltage rating of the fuse, if the fuse fails the voltage that is across your supply now also is across the both ends of the fuse, just a couple of millimeters won't hold back a high-voltage.
That is why high voltage fuses are physically larger, and have a higher burden voltage.
Navigation
[0] Message Index
[#] Next page
[*] Previous page
Go to full version