floobydust:

--- Quote from: riccardo.pittini on January 13, 2019, 07:05:45 pm ---I made the same mod to fit 5x20mm HRC fuses on a AN8009 the problem is that it is necessary to re-calibrate the multimeter as the current measurement can be affected  :)

--- End quote ---
It's only if you change the PCB and soldering path to the shunt resistors.
The fuse resistance does not affect the calibration. You have pics of your mod?

Fungus:

--- Quote from: riccardo.pittini on January 13, 2019, 07:05:45 pm ---I made the same mod to fit 5x20mm HRC fuses on a AN8009 the problem is that it is necessary to re-calibrate the multimeter as the current measurement can be affected  :)

--- End quote ---

The only thing it can affect is the burden voltage.

HKJ:

--- Quote from: Fungus on January 13, 2019, 08:59:43 pm ---
--- Quote from: riccardo.pittini on January 13, 2019, 07:05:45 pm ---I made the same mod to fit 5x20mm HRC fuses on a AN8009 the problem is that it is necessary to re-calibrate the multimeter as the current measurement can be affected  :)

--- End quote ---

The only thing it can affect is the burden voltage.

--- End quote ---

Have you seen where the sense wires for the current shunt are connected? (This can affect the 10A range, but not the uA range)

Fungus:

--- Quote from: HKJ on January 13, 2019, 09:06:17 pm ---
--- Quote from: Fungus on January 13, 2019, 08:59:43 pm ---The only thing it can affect is the burden voltage.

--- End quote ---
Have you seen where the sense wires for the current shunt are connected? (This can affect the 10A range, but not the uA range)

--- End quote ---

Burden doesn't come only from the shunt, it's the entire path through the meter from input to output.

HKJ:

--- Quote from: Fungus on January 13, 2019, 09:24:20 pm ---
--- Quote from: HKJ on January 13, 2019, 09:06:17 pm ---
--- Quote from: Fungus on January 13, 2019, 08:59:43 pm ---The only thing it can affect is the burden voltage.

--- End quote ---
Have you seen where the sense wires for the current shunt are connected? (This can affect the 10A range, but not the uA range)

--- End quote ---

Burden doesn't come only from the shunt, it's the entire path through the meter from input to output.

--- End quote ---

It has nothing with burden voltage to do, but where you connect a updated fuse to the circuit board.