Author Topic: HP 3852 calibration  (Read 2631 times)

0 Members and 1 Guest are viewing this topic.

Offline cncjerryTopic starter

  • Supporter
  • ****
  • Posts: 1308
HP 3852 calibration
« on: June 30, 2015, 05:52:33 am »
I picked-up a loaded 3852a and have been reading thru the docs, etc., playing with programming. Interesting box.  What I can't find is any information on how to calibrate either of the voltmeters, the 44701 integrating voltmeter specifically.  Any pointers?
 

Offline Julio2784

  • Newbie
  • Posts: 2
  • Country: us
Re: HP 3852 calibration
« Reply #1 on: January 18, 2016, 04:04:03 am »
  The calibration info for the HP 44701A is included as part of the HP 3852A service manual document.
(a pdf file with over 700 pages)

(This was a bit confusing for me also, there is no separate document for cal and maintenance of this plug in)

The section that you want begins in chapter 6 at 6-23 for the performance tests, and 6-35 for the actual 44701A calibration procedure.


 
 

Offline fisafisa

  • Regular Contributor
  • *
  • Posts: 106
  • Country: es
Re: HP 3852 calibration
« Reply #2 on: May 15, 2016, 09:35:54 pm »
Hi
I cannot find the service manual in question.
Have you got a link?
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29335
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: HP 3852 calibration
« Reply #3 on: May 16, 2016, 03:44:28 pm »
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 