@_RUS73_
You seem confused:
1. "an active load of 50 Ohm". Why have you written "active"?
2. "50 ohm balanced load(RF)". Why have you written "balanced"? Neither the generator output nor the scope input is balanced.
These are the difficulties of communication and understanding the essence through Google Translate:)
Now let's get to the point, and what I wanted to convey.
I got hold of a similar signal generator. The test results showed that connecting the generator via a 50 Ohm cable to the oscilloscope via a 50 Ohm pass-through load terminal to the bnc connector of the oscilloscope distorts the signal (apparently the additional capacitance of the oscilloscope input overloads the generator output). Before connecting the oscilloscope, the signal has a more correct shape. This can be easily verified by connecting a probe with a 10x divider to the circuit. The signal has a more regular shape.
Several rules for conducting their measurements are violated here:
1. The oscilloscope must introduce minimal distortion into the measured circuit.
2. The generator must have sufficient load capacity and produce the correct signal to conduct such measurements.
In this particular case, we can only correct the first point, using the probe in the 10x position to minimize the effect on the signal.
I consider the testing conducted by user BillyO to be unsuccessful due to the impossibility of observing point 2.