Products > Test Equipment
Applent AT825 calibration
<< < (3/3)
Kean:
When you say "my AT825 is showing 0.45ohm too small value", what is the resistance you are measuring?
You say they are high precision resistors, but an error of 0.45ohm could be a significant or insignificant percentage of the total value.

When I asked "What do you measure with a direct short?", I meant what was the display value?  Resistance and reactance?

As TurboTom mentions, you should use 4W measurement with a shorting bar to ensure good contact but with separate contact points for source and sense.  Doesn't the AT825 come with a shorting bar?
Veketti:
I'll update this regarding the outcome. I contacted Anbai like suggested by TurboTom and they indeed were very helpful and willing to find out the issue. Outcome was that the meter was out of calibration. They could have adjusted it but as the meter was already over seven years old, they gave good offer for new meter. So it was no brainer to buy new as freight for sending the meter to China and back would be almost half of new meter price.
Navigation
Message Index
Previous page
There was an error while thanking
Thanking...

Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod