Products > Test Equipment

ARRAY electronic loads

<< < (13/30) > >>

saturation:
You mean constant voltage?


--- Quote from: T4P on September 24, 2012, 04:24:46 am ---Well the RS232 interface cable is quite literally cheap, compared to the prices fluke charges for a rs232 interface cable  :P But yeah, one area it seriously lacks is V-set.

--- End quote ---

free_electron:
3711 Adjusted and fully in-spec now.  stupid callabs 'can't adjust' ... too lazy to open the box or download the cal manual... ah well their los, my gain.

saturation:
Great! I figured that was the case.  Another advantage for owners is that the Array's don't need much in terms of high end metrology gear either, and is easy to self maintain; one can even make small adjustments on the fly while the unit is operating to compensate for settling, due care taken naturally.

Looking forward to your progress and to LaurenceW's updates on his close case firmware based calibration his new Maynuo for comparison.
 


--- Quote from: free_electron on September 25, 2012, 04:37:47 pm ---3711 Adjusted and fully in-spec now.  stupid callabs 'can't adjust' ... too lazy to open the box or download the cal manual... ah well their los, my gain.

--- End quote ---

T4P:

--- Quote from: saturation on September 25, 2012, 03:44:35 pm ---You mean constant voltage?

--- End quote ---
Yes

saturation:
Yes, that has pros and cons, its not entirely a loss.  Since V and I are related phenomena, tests meant for CV can be done with CC for practical reasons [ that is, not following the manufacturers precise test protocol which requires CV but getting the same result ] e.g. plotting a V x I curve for a battery or PSU.  However, it may not be possible to to test NiMH chargers; if it has input protection, it will first check output voltage from the load before it switches on, so most good chargers will see the connection as open circuit or short circuit and refuse to start.


--- Quote from: T4P on September 25, 2012, 08:56:52 pm ---
--- Quote from: saturation on September 25, 2012, 03:44:35 pm ---You mean constant voltage?

--- End quote ---
Yes

--- End quote ---

Navigation

[0] Message Index

[#] Next page

[*] Previous page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod