Getting back to this, I’m still on the fence about keeping the(current) 10k resistor, and adding the external trim resistor network1.74M + 20k trimpot. Or cobble together a better 10K and a string of better resistors+ trimpot.
Back to the unit…
Since the 20Ohm range was out, yet that range uses the same current that is used in the 2 Milliohm range. Since that was good, the current source is OK. I will go into the current range trim later.
Milliohms Voltage adjust:
Set meter to 19.0 milliOhm range.
Attach leads to calibrator with output shorted/off shorted.
Adjust offset trim pot to read Zero(or as close to zero as possible).
Apply 19mV with a stable voltage source. I used the MV-106 in conjunction with the 3456A.
Use R16(trim pot) and turn it until the display reads correctly.
Set meter into Ohms range
Ohms Voltage adjust:
Disable the output to the calibrator/short the leads.
Adjust offset trim pot to read zero.
Set calibrator to read 190mV.
Use R6(trim pot) to adjust reading to be correct.
Ohm Range Current calibration.
This is pretty straight forward.
Any Range:
Attach Standard resistor(or resistor of known value) I only felt comfortable calibrating the ranges above 2000mOhms.
Null out emf voltages with offset trim pot.
Depress pulse switch.
Put meter into proper range.
Press Measure button
Repeat as necessary, adjusting appropriate trim pot on range resistor. For convenience I left the meter in continuous mode, then switched to pulse as I got closer.
Continue for every range.
Getting back to the meter…
On the display board I noticed the name Motwane, and didn’t think anything of it. I figured it was a pcb company.
Upon further disassembly I noticed the same name on the pcb’s and case. This struck me as odd.
Researching this, it is an Indian firm, still in business. That produces related T&M gear. This explains the large number of Indian companies, as parts manufacturers.
I’m now curious if Ballantine Labs designed the meter, and it was made by Motwane. Final assembly was by Ballentine labs, or what.
Questions/Comments welcome.