hey
Dave , can you shear with us the need for such accurate measurement ?
I can tell you that all of these specs are just on paper , agilent HP fluke keithley and such high end companies tend to write " max 100ppm " while when measuring 1A , the meter will be much better then 100ppm , and will show " 1.000040A " .
tell you something ? my HP34401 is not that great , but has 6.5Digit res , within 1 Day after ADJUSTMENT I get like times 10 better specs then written on the data sheet ( yes I know not to many ppl can adjust they're multimeters as often as I can _ BUT Dave has the access to Trio smart cal's instrument , hence he may be able to adjust he's 3457 )
Note , my 34401 can show spot on only on the 5,5 digit mode , while on the 6.5 it is wrong by few digits ( Ie few ppm's ) , and that's straight after adjustment
so my conclusion - use you'r meter one digit less , Ie , if it's 5.5 , use it as absolute accuracy of only 4.5 ,
although my two 3478A's are preforming absolutely amazing ! , one shows 10.0000V when the fluke 5700 set on 10.00000 ... V .... so on that meter I'll count for the fully 5.5 resolution ( ON THE DC voltage ONLY ! )