The Diff probe is lousy. (TA041)
Show us the schematic,test points you want to measure and tell us precisely why you want to measure at those points.
Quote from: nctnico on March 21, 2023, 08:56:50 pmShow us the schematic,test points you want to measure and tell us precisely why you want to measure at those points.+1
If the DUT is floating and has no ground reference the common mode voltage is poorly defined and could reach some high DC potential. Floating the scope would not necessary fix this, no would a better differential probe. You need to provide some sort of ground return for the DUT to limit the voltage it could float to.
The probe input impedance is "4 MΩ || 5.5 pF each side to ground", according to the datasheet. Isn't the resulting 4M+4M divider sufficient in order that the mean voltage between the two test points becomes the DUT's common mode floating voltage if the DUT itself is fully isolated (say 1GOhm) from ground?
Just trying to make the most of Faringdon's babble.
Quote from: Weston on March 22, 2023, 07:04:11 pmJust trying to make the most of Faringdon's babble.Unfortunately this has proven to be a lost cause time and time again.
Quote from: Kean on March 24, 2023, 01:42:19 pmQuote from: Weston on March 22, 2023, 07:04:11 pmJust trying to make the most of Faringdon's babble.Unfortunately this has proven to be a lost cause time and time again.Of course it has. I don't know why people even bother anymore..If you carefully read his first two posts, he claims that passive 100x probe and isolation transformer and TA041 show same waveform..And both must be wrong. Because he expected to see something and it showed something else... Instead of saying "oh I have oscillation here let me look in to it" he says it is not real. How does he know it is not real?Statement could have been "It shouldn't oscillate like this". Not "it's not real" how does he know "it is not real"? Because he measured twice using two different techniques and got same result... Hmmmm.... Yep, not real, I need better probe. Definition of better here being "one that will show same waveform like Ltspice predicted"..... So it's probe that is not real. LTSpice simulation is...Hence, waste of time... Because this is persistent pattern of behaviour, sadly...
By what mechanism do you think "leakage current" is causing this? The differential probe has a 4M input impedance, if you are probing a low impedance node such as the DC bus voltage it seems unlike that any current though the input impedance of the probe would cause issues. If the waveform is distorted is it possible that you are exceeding the common mode input voltage of the probe? That would distort the waveform produced by the probe. If the DUT is floating and has no ground reference the common mode voltage is poorly defined and could reach some high DC potential. Floating the scope would not necessary fix this, no would a better differential probe. You need to provide some sort of ground return for the DUT to limit the voltage it could float to.
If an ordinary differential probe is not good enough, what about optically isolated probes?Seems they support common mode voltages of tens of kV, and CMRR of 150dB and more Price is scaring, though