I still couldn't see Fluke printing 1000V MAX on the front of all the 8840A/AF units, where the AC board is not optional, for military customers. So that got me thinking more about this.
I went through the True RMS theory of operation on page 5-19 of the 8840A manual. The block diagram shows how the input is scaled based on the selected range. In all cases, the output from the TRMS board is always 0-2V to the track/hold circuit:
- 200mV, r1: divide by 5, multiply by 5, multiply by 10 = 2V
- 2V, r2: divide by 5, multiply by 5 = 2V
- 20V, r3: divide by 5, multiply by 0.5 = 2V
- 200V, r4: divide by 500, multiply by 5 = 2V
- 2000V, r5: divide by 500, multiply by 0.5 = 2V
It also states that the input is clamped by MOVs to 1560V. So 1000V RMS (1414V) should be no problem, even for the 8840A and 8842A. Of course different firmware would be needed, and I suspect some of the input protection is different on the 8840A/AF since 1560V might be a bit close to 1414V.
With regard to the calibration discussion, I don't see how you could rely on anything without a somewhat recent calibration (self or otherwise). That is the entire point, provide evidence that it works correctly so you CAN rely on it. In the context of this thread, the OP is not going to have the necessary equipment to calibrate it himself, so that means paying for it.