Author Topic: How to test a RF Gen board out of circuit  (Read 1484 times)
Offline NY2KWTopic starter
« on: November 08, 2019, 02:29:20 pm »
I have a R&S SM300 RF board that appears to have a bad HMC307 digital attenuator.  There are 4 of them that control the RF level output (-125 to +13 dBm) and each can attenuate in 1 dBm steps from 0 - 31 dBm.   Service manual is useless and no schematics available (i called R&S) Picture is attached.  The HMC307 has control lines that are controlled by (-5v).   The HMC307 datasheet is also attached.
I know what has to be done - tie in Vee and a GND then test each control line while providing a input signal and look at the output on my SA.  The issue for me is the space is so damn tight.
Any advice appreciated
Jerry NY2KW
* HMC307.pdf (207.97 kB - downloaded 82 times.)
Re: How to test a RF Gen board out of circuit
« Reply #1 on: November 08, 2019, 02:30:14 pm »
sorry here is the picture
Online PA0PBZ
« Reply #2 on: November 08, 2019, 03:03:23 pm »
Are they controlled from the same board or from somewhere else? If it is from the same board you can probably not override the lines without the chance damaging the controlling stuff? On the other hand, if they are controlled from outside you could maybe pickup the control lines...
« Reply #3 on: November 08, 2019, 03:45:00 pm »
It has a PC control board (Linux) for display and remote.  The generator is made of 2 boards, an analog and digital.  No schematics so hard to say where the control circuitry is located but it is probably located on the PC control card or the generator digital board.  The attenuators are on the analog generator board.   Any advice how to work in such tight space around each HMC307?
Online PA0PBZ
« Reply #4 on: November 08, 2019, 04:09:35 pm »
There is some kind of controller IC (D8) on the analog board:
868608-0
The control lines for D8 are coming from the digital board:
868612-1
Maybe it is possible to use the control lines if you are able to find out how they work?
"Service manual" is here: https://elektrotanya.com/rohde_schwarz_rs_sm300_vector_signal_generator_sm.pdf/download.html
Offline chick0n
« Reply #5 on: July 06, 2021, 08:17:38 am »
I built an Extender Cable for the RF-Board, out of this:
https://www.amazon.de/gp/product/B071794QXD/ref=ppx_yo_dt_b_asin_title_o00_s00?ie=UTF8&psc=1
Works great. You have to cut it for 30Pins.
The following users thanked this post: TurboTom, analogRF