Products => Test Equipment => Topic started by: tombo on April 30, 2023, 12:18:44 pm
-
Hi All
new to the forum, been using my BK precision Multimeter to measure mains voltage on a hot water boiler
here in the UK it should read 240v but is telling me 150v
Ivegot a 20 year old no name cheapy meter that reads the voltage correctly
any ideas, is if faulty or am I using it wrong?
Cheers
-
My guess is something wrong with the meter but to be sure I think it's helpful if you post the model number and also what settings you use.
-
Hi,
Thanks for the reply its a BK precision 2712, irritating if its faulty as these are quite expensive though not fluke costs
please see photo of meter measuring mains voltage, you can see the model and the reading I am getting
Cheers
-
The meter settings looks correct. So it could be a faulty meter or maybe just faulty cables. Due to capacitive coupling one may still get some AC votlage, though lower than it should be.
One could do a test in ohm mode to see if it is reading Ok there. The ohms part is more sensitive to a bad cable.
-
I see only one probe and it is on the neutral pin. Where is the other probe?
-
sorry bad photo other lead is behind on the live terminal
I'm guessing the meter is faulty >:( ive just used that same leads in my cheapy meter and it reads correctly see photo
thanks for the help ill see if there is a way to get it repaired
-
Try moving the selector...