Author Topic: Newport 3040 temperature controller.  (Read 906 times)
Offline jchw4Topic starter
« on: July 05, 2023, 07:47:31 am »
I got a Newport 3040 temperature controller and I have trouble setting it up.
I am trying to make it work with a cheap car thermoelectric cooler, which I rewired directly to the controller.
The issue is that it does not stabilize. And I don't have experience with thermoelectric controllers. I need some help understanding whether it works as intended or something is broken.
I found user manual here: http://www.iproweb.fr/test/0%20-%20MATOS%20GUILLAUME/NEWPORT%20-%203040/DOC%20US.PDF
3.4.2.4 Gain
The Gain function controls two parameters of the hybrid PI control loop;
proportional gain and integration time.
When the actual temperature and the set point are different, an error voltage is
generated. This error voltage is directly related to the difference in the actual and set
point temperatures. The error voltage is then amplified by the proportional gain.
This amplified error voltage controls the amount of current driven through the TEC.
The higher the gain, the more current will be driven for any given temperature
difference, with the maximum current being determined by the current limit.
The error voltage also drives an integrator. The integrator’s output also controls the
amount of current being driven through the TEC. The integrator is an amplifier
whose gain is proportional to time. The longer a given error voltage is present, the
more current will be driven through the TEC, with the maximum current being
determined by the current limit. The speed at which the integrator’s output increases
is the integration time, which can be “Slow” or “Fast”.
The allowed Gain values are: 0.2 Slow, 0.6 Slow, 1 Slow, 1 Fast, 2 Slow, 3 Fast, 5
Fast, 6 Slow, 10 Slow, 10 Fast, 20 Slow, 30 Fast, 50 Fast, 60 Slow, 100 Fast or 300
Fast. The number actually defines the proportional loop gain. The slow/fast suffix
indicates the speed at which the integrator’s output increases. The slow setting allows
for larger masses or greater distance between the sensor and the thermo-electric
cooler by slowing the speed of the integrator.
Both the proportional gain and the integration time must be matched to the thermal
characteristics of the TE cooler and sensor. If the settings are incorrect, the
temperature set point will take an excessive amount of time to settle, or it will
oscillate around the set point and never settle.
The Gain setting depends on the type of TE cooler that you are using, but we can
suggest guidelines for selecting the proper gain. Set the gain to 1 fast and increase it
until the actual temperature oscillates around the set temperature. Then reduce the
gain to the next lower value.
The in Principles of operation on page 32-33 (42-43 in PDF) they say:
4.2.7 Proportional Amplifier and Integrator
The proportional amplifier is part of a digitally controlled gain stage consisting of the
analog switches and their associated resistors. The analog switches vary the ratio of
resistance in the feedback circuit to change the gain.
The signal from the difference amplifier is sent to an integrator which reduces the
difference between the set point temperature and the actual temperature to zero,
regardless of the gain setting. An analog switch discharges the integrating capacitor
whenever integration is not required to prevent unnecessary difference signal
integration.
So my guess is that it is actually a digitally controlled fully analog device.
I.e. once the mode is set, the rest is done on the analog side. Digital side at this stage
only does monitoring.
I set up a script to try all the 16 modes for 5 minutes each.
Room temperature is about 25C, target temperature 30C.
Temperature and TE current values are from the controller.
I just checked that it measures current correctly. So the values seem to be real.
Here is what I got:
What surprised me is that it continues to drive the TE at nearly the full current until after target temperature is reached and then tries to go in the opposite direction at full power. Here is the central part of the graph above:
The “10” mode is interesting because it offers both “slow” and “fast” integrator modes. So we can compare them:
Actually I can see that there was one case, "60 Slow" which has decreasing amplitude. But I don't feel that the stabilization time is usable:
So my question is whether this device is working.  And if not, what could be wrong with it?
Upd: I had to re-upload all the images to make img tags work.  :-\
« Last Edit: July 05, 2023, 08:21:38 am by jchw4 »
Offline Kleinstein
Re: Newport 3040 temperature controller.
« Reply #1 on: July 05, 2023, 08:36:11 am »
The curves look like a poorly adjusted regulator that oscillates or is very close to oscillation. To me this looks like the instrument itself works, but the system used for testing may be not suitable for this regulator.  The settings seem to be limited to a small number of settings and as it looks as PI only (so no PID).
A significant point is how fast the temperature sensor reacts to power changes. The larger the delay, the more difficult to regulate is the system. One may still be able to use the controller with the temperature sensor paced closer the peltier element, so that it reacts faster.
The following users thanked this post: jchw4
« Reply #2 on: July 05, 2023, 09:49:59 am »
The sensor is attached to the screw holding the element:
But the issue is that the temperature values are read from the controller itself. So this is how it sees the temperature. If I attach the sensor even closer the controller reaction delay will still be the same.
Let's look at the default mode "30 Fast" that is suggested as a starting point in the manual:
The zoomed-in version:
So you can see that it continues heating or cooling at full speed even  when it already knows that the sensor has reached the target temperature.
« Last Edit: July 05, 2023, 09:51:40 am by jchw4 »
Online ch_scr
« Reply #3 on: July 05, 2023, 10:48:08 am »
You could try lowering the gain of the system by setting a lower max current for the TEC. Since the control loop is analog, with digitally controlled switched elements (R, C), one would have modify the hardware to make it work.
Offline Kleinstein
« Reply #4 on: July 05, 2023, 11:06:47 am »
The behavior is normal. The time the heater is still on after reaching the set point is actually relatively short. This suggests that there is some anti-windup active. This is common with a digital regulator and a bit rare with an analog regulator. So chances are the regulator is digital.
The position of the sensor aready looks not that bad, though the liquid in direct contact to the sensor will slow the response. It would not be ideal for an accurate control, but some isulation around the actual sensor could help.
« Reply #5 on: July 05, 2023, 11:36:51 am »
I tried this manually (so that I don't have any data) but I think that it did not help.
I'll repeat my experiment with lower current limits in a few days.
« Reply #6 on: July 05, 2023, 11:40:00 am »
I feel that adding more thermal mass to the target side of the TE could help. But still, looking at the device specification:
I don't understahd how short term stability of 0.0005C is expected to work.