With what you said, manupthehills in your last post, I made measurements with the following configuration.
An HF generator connected to a single channel of the DS1102Z-E with its trigger output connected to a second oscilloscope.
I measured a refresh of the DS1102Z-E of 30000wfms/s (period of 33µs) obtained ONLY when:
- only 1 oscilloscope channel is used
- frequency of the signal generator lower than 10Mhz (approximately)
- oscilloscope time base is equal to 50ns
These conditions for having the highest refresh rate are certainly different from one oscilloscope to another and from one manufacturer to another and especially the purchase price.
Regarding the brightness problem, I measured under the same conditions as my first post:
DS1102Z-E time Base | Trigger (Time between 2 wfms) |
2ns | 125µs |
5ns | 120µs |
10ns | 57µs |
20ns | 76µs |
50ns | 87µs |
100ns | 150µs |
200ns | 290µs |
The low luminosity of waveform on the screen of the oscilloscope with a time base of 5ns cannot be explained with the refresh rate (wfms/s) of 120µs. This value is higher with other time base values and the luminosity is normal.
As you said, manupthehills, it may be due to a bug in the software or, in my opinion, a hardware problem with the motherboard...
You, the other users of a DS1000Z-E, what is the version of the motherboard?