Caliaxy:

--- Quote from: Neutrion on January 24, 2022, 07:24:50 pm ---The 87V has no mV scale so if we are talking about the V AC scale the Brymens read also correct on the v AC scale.

--- End quote ---

That's actually the interesting thing we noticed while having this discussion in the BM789 thread.

Fluke 87v has a DC mV position on the dial (with a single 600mV range). There is no AC mV position on the rotary dial per se, but there is a 600mV AC range on the AC V position of the rotary dial. I suspect they did this precisely to avoid the issue under discussion here without adding an extra position on the dial (the alternative would have been having a common AC/DC mV position on the rotary dial and switch between AC and DC with a soft touch button).
bdunham7:

--- Quote from: Neutrion on January 24, 2022, 07:24:50 pm ---The 87V has no mV scale so if we are talking about the V AC scale the Brymens read also correct on the v AC scale.

--- End quote ---

It actually does, in fact IIRC in Hi-res mode it has a 200.00mVAC range, so 10uV resolution. 
Caliaxy:
A quick fix for the OP (other than a capacitor) would be to use a lab power supply to subtract the offset. Set the output of the power supply to the offset voltage (7.5V), connect (V-) to the GND of the DUT and use BM689S to measure the AC voltage between the measuring point and (V+) on the power supply (as opposed to GND or V-).
Fungus:

--- Quote from: bdunham7 on January 24, 2022, 07:50:06 pm ---
--- Quote from: Neutrion on January 24, 2022, 07:24:50 pm ---The 87V has no mV scale so if we are talking about the V AC scale the Brymens read also correct on the v AC scale.

--- End quote ---

It actually does, in fact IIRC in Hi-res mode it has a 200.00mVAC range, so 10uV resolution.

--- End quote ---

So the OP can simply use AC range (or AC+DC range) with no problems?
Neutrion:

--- Quote from: bdunham7 on January 24, 2022, 07:50:06 pm ---
--- Quote from: Neutrion on January 24, 2022, 07:24:50 pm ---The 87V has no mV scale so if we are talking about the V AC scale the Brymens read also correct on the v AC scale.

--- End quote ---


It actually does, in fact IIRC in Hi-res mode it has a 200.00mVAC range, so 10uV resolution.

--- End quote ---

Maybe this one goes to the  87V but it would be still interesting to compare the 87Vs 200,00mVAC range(0,7%+1d) with the 789s 6,0000VAC range   0,5%+30d in an actual test.

Did by the way anyone tested this scale of the 87V with a DC bias, and for overrange indication?