| Brymen BM257 calibration error 'C_Er' |
| (1/2) > >> |
| Gabri74:
Just happened the same problem as describe in another old thread: https://www.eevblog.com/forum/testgear/brymen-257s-lost-calibration-data/ Today to my BM257 had a sudden low-battery warning, display flickers, meter power off itself and the the 'C_Er' message is displayed :'( Tried entering/exiting calibration mode procedure to reset the error as suggested in the above thread without success. Can anyone suggest where to send it to calibrate from Italy ? Moreover, will it be worth to calibrate it ? I don't know how much could it cost, but the meter is several years old and if it's too expensive I'll probably buy a new one. Thanks in advance |
| J-R:
What if you do the first calibration step for resistance (short), then power cycle? |
| Gabri74:
Thanks for your suggestion. Followed this procedure: 1) Turn Rotary Knob to "OFF" position. 2) Press and hold "SELECT" and "HOLD" buttons, and then turn "Rotary Switch" from "OFF" to "" function position within one second. 3) "000.0" is displayed 4) Input 000.0, press "SELECT" and then "REL" buttons simultaneously 5) Wait till reading is stable, press "SELECT" and then “HOLD” buttons simultaneously. 6) “300.0” is displayed I had to do it several times, because on the first tries at point 6) (after pressing select+hold) it was showing and empty display with the 'nF' symbol only instead of 300.0 ohm. After several retries it worked, but instead of going to 300.0 as shown in point 6), it went straight to 30.000K... And now my ohm range is broken: when shorting, it shows 1.000K and 1.500K intermittently :'( I think I screwed up the calibration completely... |
| mwb1100:
welectron offers calibration for Brymen meters for 60 EUR + VAT + shipping. (BM780/BM850s/BM860s is 10 or 15 EUR more) They sell BM257s meters for 90 Eur + VAT + shipping (I think shipping is 10 EUR within the EU), so the cost of calibration is pretty close to getting a new meter (since you'd have to pay to ship your meter to them). If you do order from them, ask them for the EEVBlog discount (5% I think). |
| J-R:
I think you just went too far in the procedure and entered the 300 Ohm calibration point. Can you find a 300 Ohm resistor to fix that mode? Or are you saying it doesn't ever show the 300 Ohm calibration point any more? If so, it does sound like it has bad flash memory, so it would have been a lost cause anyway. |