I recently purchased a BM867 based on the EEVBlog video overview, comments on the forum and some other online reviews. It's a fantastic meter and I get a lot of joy using it.
However, I may have discovered a bug on it and would like to know if others could replicate it or if it is a problem specific to my unit. It has to do with Crest detect Mode in the AMP range.
The test is simple:
- short the common and 10A input jacks with a test lead
- switch to the A scale
- now press CREST button
- does yours also measure at least 0.65A as a max? (without having any current flow through it)
Can owners of the BM869 perhaps try the same test?
The above problem does not allow me to take current in-rush measurements!
Update from Brymen:Thanks for using our products and feeding back comment to us.
This is not its bug. Actually it is its nature out of its design. BM869 Crest mode design is targeted to capture the peak value short to 0.8ms in duration. To hit this target, very high speed sample rate is in need. In such case, its noise rejection can not be good as normal measurement mode naturally. That is the reason why we add an additional tolerance for CREST mode to manual specs as followings.
Crest mode (Instantaneous Peak Hold)
Resolution: 5000 counts
Accuracy: Specified accuracy +- 100 digits for changes > 0.8ms in duration
--deckert