Hello,
i just did a quick test with my BM867s and a SDG1032X. Below is the result. I don't know whether the deviation is caused by generator or multimeter since my second multimeter is not here at the moment.
SDG1032X was set to 2,828V peak-peak in order to get approx 1V rms. BM867s was in auto range but remained in 5V range during all measurements.
Will test again when my Fluke multimeter is back to compare results.
Thanks for efforts and sharing,
Can you perform next measurements at 2Vpp - then we can easily compare them to other test results? - My & 2N3055
I observe the following issues
- BM867s shows also the frequency 1 digit lower e.g x.99KHz instead of (x+1),00KHz -- same as BM869s
- huge difference in 0..100KHz range between BM869s/Fluke187 and BM867s +-0,8% vs +-4%
- already at 5KHz BM867s is 3% off - It is still within Brymen specs, but this is definitely showstopper for me.
- Why BM867s characteristics goes up at MHz range? - what type of cable do you use, Do you terminate the coax cable at the end?
If the next BM867s measurement confirms the same characteristics - I'll byte the bullet and buy BM869s