Hi,
On my BM 869, I had a closer look on 3 separate circuit aspects of this effect:
- ranging method
- stability of the Ohm measurement function
- over range detection
I also see this annoying effect, even in my basement lab, which is relatively quiet concerning EMI, and even w/o any provoked AC mains signal injection.
Down-ranging always occurs at a fixed / digital count of about < 5000 counts.
Up-ranging occurs at random counts. In DCV mode, that's about > 53125 counts, in Ohm modes it's about > 53300 counts.
In the first place, I tripped about other contributions here, which reported >56000 counts, or so.
Up-ranging is therefore initiated by an over range condition, not by a fixed upper digital A/D value, like in other DMMs (e.g. 1 200 000 ... counts for all HPAK bench DMMs).
Obviously that over-range detection inside the BM869 is realized analogously by a window comparator (OpAmp) which probably monitors directly the input signal over an additional input resistor.
In fixed 500kOhm range, as far as no over range occurs, the measured values between 500 and 530kOhm are always quite stable, due to the averaging characteristics of the multislope A/D. Of course, for any DMM with such a low sensing current of 1µA, you can easily provoke this reading to fluctuate by injecting large AC signals. Latter I would regard as being completely normal, and not being a 'Resistance Quirk' at all.
Approaching these 53300 counts, the BM869 quickly switches between OL and a mostly correct measurement value.
So the culprit lies only in this analogue overload comparator, which reacts much too sensitively on induced EMI signals.
I assume, that the additional capacitor simply gives a low pass RC filter to suppress that effect, making OL detection slower, but would probably not greatly affect the measurement.
Btw.: We have discussed such OL detection problems as well on the 121GW for AC+DC signals, but the other way round.
There, the OL detection is done digitally only, which leads to an undetected OL condition and a saturation of the input buffer, and in turn to (partly dangerous) false readings.
There, an additional analogue OL detection would have solved the problem.
So in the end, this is a weak design on a very limited use case, but it does not greatly affect the usability of this DMM.. simply switch to manual ranging under this condition.
Frank