Products > Test Equipment
Brymen Crest and Min/Max Function VDC Capability
armandine2:
Fluke 287 on min/max - just tested the battery on start - 3 times
first min reading was negative, under a volt - not sure what that was about? - ghost voltage, as I set min/max prior to attaching leads to battery :palm:
second reading min was over 10 volts
third reading - saved
I'm thinking may be my 287 isn't so quick as other's 289?
Ed to add Peak info from manual
chebo:
Let's raise a forgotten topic... I couldn't find a newer topic about this Crest mode.
I recently purchased a BM789 and decided to use it to check my car battery. And at the same time I compared it with what the oscilloscope shows. This is what happened.
The result is more or less the same, but there is still a small difference.
Then I decided to conduct a small experiment at home with this scheme.
I manually closed and opened the switch shown in the diagram, simulating a voltage drop. The average value for several measurements in the Crest mode was 7.60 V. As a result, the relative error was approximately 4%.
Navigation
[0] Message Index
[*] Previous page
Go to full version