Author Topic: Brymen, difference in reading between mA and uA scale  (Read 5205 times)
Offline MarioBros69Topic starter
« on: August 13, 2023, 08:15:55 am »
These days I have used a Brymen BM869s, an impressively robust multimeter
But I have found the following, I do not know if it is my ignorance or a problem with my unit
I put it on the mA scale and measure the current through a 2k resistor connected to a DC 10v source, this should give 5mA, the measurement on the mA scale is very accurate 4.999 mA
The accuracy stated in the specifications for the mA scale is
50,000mA 0.15%+20d 3.3mV/mA
Now I turn the selector and change to the uA scale (micro amps) and the measurement gives a value of 4670.4 uA
We have a difference of 230 uA between the two scales for the same circuit. If I didn't know the current of the circuit I'm measuring, I wouldn't know which of the two is correct.
The precision stated in the specifications for the uA scale is
5000.0 0.1%+20d 0.15mV /uA
I do a little math and the result is that for the uA scale and for this measurement the error can be plus or minus 6 uA but not 230
I am also aware that the internal resistance (Shunt) according to the specifications is 3.3 ohms for the mA scale and 150 for the uA scale.
My question is, should the two scales show the same measurement value or is it correct that they appear differently because the internal resistance is different on each scale?
Could someone with a fluke or a Uni-t perform this simple test?
« Last Edit: August 13, 2023, 02:15:43 pm by MarioBros69 »
Offline Performa01
  • Country: at
Re: Brymen, difference in reading between mA and yA scale
« Reply #1 on: August 13, 2023, 08:44:50 am »
Now I turn the selector and change to the yA scale (micro amps) and the measurement gives a value of 4670.4 yA
The precision stated in the specifications for the yA scale is
5000.0 0.1%+20d 0.15mV /yA
I am also aware that the internal resistance (Shunt) according to the specifications is 3.3 ohms for the mA scale and 150 for the yA scale.
Of course the measurements will differ, thats why Daves µCurrent exists - to lower the burden voltage for current measurements.
The instrument measures current. You are not using a constant current source but a voltage source together with a resistor instead, both with a tolerance most probably exceeding the one of the multimeter.
Ohms law says: I = U/R:
10 V / (2000 + 3.3) ohms = 4.99176359 mA
10 V / (2000 + 150) ohms = 4.65116279 mA
Seems like your measurement results are reasonably plausible.
Offline Fungus
« Reply #2 on: August 13, 2023, 09:04:48 am »
No.
Yes.
That's the entire point of having two different switch positions instead of just autoranging. Each position has an internal resistance that's optimized for the range of currents that will be read there and so have the minimum influence on the circuit being measured.
Offline J-R
« Reply #3 on: August 13, 2023, 09:55:39 am »
Side note, double-check your specific DMM yourself and don't assume the manual is perfect.  Resistances on my BM869s:
A - 0.0276 Ohms
mA - 1.758 Ohms
uA - 101.543 Ohms
And my BM789 is 102 Ohms for uA, not 200 Ohms as stated in the manual.
Offline BILLPOD
« Reply #4 on: August 13, 2023, 01:22:45 pm »
I thought a yA was a Yactoampere :scared:
Offline Shonky
« Reply #5 on: August 13, 2023, 02:04:26 pm »
yoctoamp
OP: just use a lower case u in place of mu for micro. Lower case y = 10^-24 not 10^-6
« Reply #6 on: August 13, 2023, 02:17:13 pm »
yoctoamp
I'm sorry, I have already corrected it.
« Reply #7 on: August 13, 2023, 02:22:47 pm »
The truth is that I don't understand this very well..., if the measurements differ, which of the two is correct, the one on the mA scale or the one on the uA scale?
How do we know then the current that circulates through the circuit when we are not measuring?
Offline sonpul
  • Country: ua
Re: Brymen, difference in reading between mA and uA scale
« Reply #8 on: August 13, 2023, 02:47:07 pm »
The current that you see on the display is real and truthful.
If you set 10V 5mA with a laboratory power supply, then without a resistor you will see exactly 5mA at any position of the mA or uA switch.
Carefully! if you make a mistake, you can damage an expensive fuse. Just believe.
« Last Edit: August 13, 2023, 02:50:13 pm by sonpul »
Offline IanB
« Reply #9 on: August 13, 2023, 03:32:41 pm »
Or even better, use the actual µ character for µA.
Both are correct. The truth is that inserting a current meter into a circuit will affect the current. So such measurements have to be taken with care.
With difficulty. One possibility is to find an existing resistor in the circuit and measure the voltage drop across it. So in your case, you have a 2 kΩ resistor. Measure the voltage across that and use Ohm's law to find the current. For increased accuracy, measure the resistance of the resistor out of circuit so you have a better estimate of its value.
Another possibility is to use a higher current scale on the meter, such as the mA scale or even the A scale. Or use a device like Dave's µCurrent.
Online BeBuLamar
« Reply #10 on: August 13, 2023, 03:34:25 pm »
Both readings are correct. They indicate the actual current that flow thru the circuit but neither is the current flowing thru the circuit without the meter. The meter causes the current to flow slightly less than if there is no meter.
The following users thanked this post: Smokey
Offline TimFox
« Reply #11 on: August 13, 2023, 04:32:15 pm »
Note that the "burden" resistance, which is the actual resistance between the two ammeter terminals, is higher than the "shunt" resistance, where the voltage due to the applied current is measured.
(Not necessarily a large difference, note J-R's measurements above.)
In your circuit, the total resistance seen by the voltage source is the sum of your external resistor plus the burden resistance.
« Reply #12 on: August 13, 2023, 05:58:10 pm »
Shouldn't the tester compensate the measurement based on its internal resistance to show the actual current flowing through the circuit?
Something like..., if my internal resistance on the uA scale is 100 ohms I add 200 uA to give the actual current going through the circuit when it is not being measured.
Offline wasedadoc
« Reply #13 on: August 13, 2023, 06:40:08 pm »
It is impossible for a meter to calculate the required correction because the size of the "correction" depends on the impedance and voltage of the source of the current.  For example both a 10 Volt source with 10k resistor and a 1 Volt source with a 1k resistor would produce 1 mA current when the meter is not in circuit.  Inserting a meter with 100 Ohm shunt has a different and larger effect in the 1V and 1k case. The meter cannot know which of the two cases or any of the infinite number of other possibilities (X Volts and X Ohms) to compensate for.
« Last Edit: August 13, 2023, 06:45:06 pm by wasedadoc »
« Reply #14 on: August 13, 2023, 07:58:55 pm »
Thanks for the explanation, now I understand.
What is paradoxical is that it is the expensive meters that come with the uA scale and we find that the uA scale of a meter is generally less accurate than the mA scale since its internal resistance is usually higher.
Offline TimFox
« Reply #15 on: August 13, 2023, 08:42:26 pm »
You still don’t understand.
In a battery-resistor-meter circuit, the burden resistance of the meter reduces the actual current into the meter.
What you measure is what is actually flowing, not what would flow into a short circuit.
« Reply #16 on: August 13, 2023, 09:57:23 pm »
Yes, I think I understand it, when making the measurement I have added a resistor to the circuit (which is the internal resistance of the meter) therefore the measured current is lower than the real current
Offline TimFox
« Reply #17 on: August 13, 2023, 10:18:35 pm »
What do you mean by "real current"?
Your meter is reading the actual current through its terminals, which is lower than that which would flow into a short circuit, since your meter is not a short circuit.
Offline siealex
  • Country: ua
« Reply #18 on: August 13, 2023, 11:18:44 pm »
You haven't seen a cheapo DT-830B (or another 7106-meter) with 1 kΩ input resistance on the 200 μA range.
Offline J-R
« Reply #19 on: August 14, 2023, 01:00:17 am »
To be clear, typical bench and handheld DMMs don't technically measure current directly, they measure the voltage drop across a known precise resistance (shunt) and calculate the current using Ohm's law.  The reason the burden voltage is a "problem" is because these DMMs have a limit to how low of a voltage they can detect.  Since V=IR, the resistance must be made larger if the current is smaller in order to maintain a voltage range that can be detected/measured.  The DMM manufacturers make a design choice for each current range depending on the capability of the DMM.  That is why lower-end DMMs typically have a higher burden voltage.  (The 121GW uses an amplifier in some ranges similar to what the uCurrent does in order to reduce the burden voltage as much as possible.)
The absolute minimum voltage the BM869s can detect is 0.001mV (500,000 count mode).  The lowest current it can measure is 0.001mA and 0.01uA.  Based on the shunt values I provided earlier, we can see that 0.001mA results in a value of 0.001758mV and 0.01uA results in a value of 0.00101543mV.  So both of these are clearly chosen to line up with the best the BM869s can do, 0.001mV.  (10A shunt calculates to 0.00276mV.)
Another way to measure current is with a Hall Effect sensor found in DC clamp meters, but most are going to bottom out in the low mA area.  Forget uA or nA.
Speaking of mA/uA/nA, I'll maintain there is no huge reason to need the uCurrent or similar since you can simply use your own known resistor as the shunt.  For example, the uCurrent uses a 10k resistor for the nA range, so assuming you need to measure 1nA, the voltage drop across your shunt would be 0.01mV which the BM869s can manage well enough.  It becomes trivial at say 50nA, which is 0.5mV.
Offline Monkeh
« Reply #20 on: August 14, 2023, 01:19:06 am »
No, the measured current is the real current. The current without the additional resistance in circuit is different.
Offline EEVblog
« Reply #21 on: August 14, 2023, 05:34:54 am »
Both are correct.
The act of putting your current meter in series changes (in this specific case) the currect because the meter has a shunt resistor in it.
You always have to know what you meter is doing to your circuit and whether or not it is going to change the circuit current or voltage.
« Reply #22 on: August 14, 2023, 08:46:50 am »
Maybe I expressed myself wrong, by real current I meant the current that flows through the circuit when I am not measuring
« Reply #23 on: August 14, 2023, 08:48:46 am »
In your previous post I see that you have both the BM869s and the BM789 according to your point of view, which of the two is more accurate when measuring current?
Thank you for this excellent clarification
« Last Edit: August 14, 2023, 09:05:59 am by MarioBros69 »
Offline Fungus
« Reply #24 on: August 14, 2023, 10:06:17 am »
Do you mean which has the lowest burden voltage?
It's specified in the manuals/datasheets of those meters.
(One of them is better on the uA range and the other is better on mA and A - you can't win!   :)  )
« Last Edit: August 14, 2023, 10:13:09 am by Fungus »