Products > Test Equipment
Can I trust this cal cert
<< < (4/4)
J-R:
For sure don't expect the 8.5 digit mode on the K2001 to be comparable to a true 8.5 digit unit, but I tested it out with various DCV references that I had calibration data for and found it better than expected.  You want maximum NPLC as well as advanced filtering if I recall correctly.  As with most DMMs with such a feature, one of the primary areas I found it works well is at the lower end of the range.  So don't expect too much at 1.9V for the 2V range for example, but it can be great at 3V in the 20V range where you'd otherwise lose a whole digit.

When performing such operations you really need the calibration data and you need to reproduce the calibration conditions.  Then these older units can do more than what is expected, IMHO.
Navigation
Message Index
Previous page
There was an error while thanking
Thanking...

Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod