Author Topic: 34401A different readings using GPIB  (Read 1562 times)
Offline Penguin36Topic starter
« on: March 09, 2018, 08:32:29 am »
I get different readings measuring a LM399AH based voltage reference when using GPIB on my 34401A.
Using the 34401A after Reset and switching to 6 digits, the display shows stable 5.02233V (thats my reference voltage) for hours.
As I understand thats '6 digits fast' with 10 PLC, right?
When using GPIB with raspberry pi, I get higher values for the same voltage reference.
GPIB settings are:
CONFigure:VOLTage:DC 10,MAXimum
SENSe:VOLTage:DC:RANGe 10
SENSe:VOLTage:DC:NPLCycles 10
and than READ? and write to a logfile.
Logfile shows 8 digits after comma and values are higher than using 34401A without GPIB.
Values are 5.02242***  (*** different numbers because of noise I think)
Resetting the 34401A as before, don't use GPIB, I get the lower voltage again (5.02233).
Why do I get different readings?
« Last Edit: March 09, 2018, 08:36:07 am by Penguin36 »
Online HighVoltage
Re: 34401A different readings using GPIB
« Reply #1 on: March 09, 2018, 09:55:40 am »
Because you have a "MAXimum" in your read statement
If you would set the instrument manually to MAX you most likely will also get the higher number on the display.
« Reply #2 on: March 09, 2018, 10:11:08 am »
The MAXimum statement in CONFigure:VOLTage:DC 10,MAXimum means the resulotion, so why should this influence the measurement and show higher values?
Online HighVoltage
« Reply #3 on: March 09, 2018, 10:58:40 am »
Oh, sorry, I was wrong!
I never used the MAX command there, I always put the number in that I want the resolution to be.
What happens, if you use:
CONF:VOLT:DC 10,0.00001
Offline Andy Watson
« Reply #4 on: March 09, 2018, 11:41:16 am »
What happens if you insert a delay between issuing the read command and then recovering the value?
« Reply #5 on: March 09, 2018, 11:49:28 am »
I‘m on work now. I will try your settings later and post results.
« Reply #6 on: March 09, 2018, 06:40:10 pm »
With CONF:VOLT:DC 10,0.00001 the readings are correct now. I don‘t know whats wrong with the previous settings. Anyway ...
Online HighVoltage
« Reply #7 on: March 10, 2018, 10:53:04 am »
Great to hear, that you got it to work.