Author Topic: Can't get my Rigol 1054z to trigger?  (Read 3264 times)
Offline LektroiDTopic starter
  • If it didn't explode, I'm happy.
    • Music here
« on: April 01, 2015, 10:02:07 am »
I posted before about having trigger issues, I can't find the original thread now, but I still can't seem to get it to trigger. I've attached the 4 probes to the calibration posts, and this is what I get. I have tried loads of different trigger options, with different settings, but I cannot get the waveforms to lock into place.
I am new to DSOs, and my analogue scope would lock straight on from any channel without a problem, but this thing is jumping around all over the place, it's unusable like this.
Is this somthing I've missed, or is the scope duff?
Offline JohnnyBerg
Re: Can't get my Rigol 1054z to trigger?
« Reply #1 on: April 01, 2015, 10:03:58 am »
Perhaps increase the trigger level?
As far as I can see, you're triggering on CH1, but the level is at the low level of the square wave.
« Reply #2 on: April 01, 2015, 10:07:56 am »
+1
Look at the Trigger setting at top right, it shows yellow (CH1) and zero volts.
Raise the level.
Offline electr_peter
  • Country: lt
« Reply #3 on: April 01, 2015, 10:42:32 am »
Trigger is not set Edge type. Change trigger type.
Offline Solder_Junkie
« Reply #4 on: April 01, 2015, 12:22:43 pm »
Use one probe connected to CH1, connect probe to probe test terminals, turn off CH2, 3 and 4. Press "AUTO".
Does the scope display a steady waveform? It should, if not there may be a problem.
As others mentioned, your triggering point is too low in the images shown.
Offline Fungus
« Reply #5 on: April 01, 2015, 02:57:33 pm »
Your analogue scope probably triggered in AC mode, this one doesn't.
Look at your image: You see the orange 'T' on the right hand side? That's the trigger level. You need to move it up a bit because right now it's in the 'noise' at the bottom of the signal.
« Last Edit: April 01, 2015, 03:03:07 pm by Fungus »
Offline aveekbh
  • Country: in
« Reply #6 on: April 01, 2015, 03:34:25 pm »
Select Edge Trigger (from the trigger menu, button under the trigger knob) and raise the trigger level. That's the most basic trigger, and it should work.