Let me add some additional clarity to my last post, because the values I picked as examples (20 deg and 10 deg) could lead to ambiguity...so I'll pick a different example.
1) Insert probe into slot 1 (before any adjustments)
2) The display shows 20.0 degrees ambient (example for illustration).
3) Enter the "offset mode" by pressing down the two keys
4) The left display shows 20.0 degrees (actual) and the right display also shows 20.0 degrees. They call this an "offset" reading, but it's not really an offset in absolute terms. It's the same measurement as the last ambient reading. This is important...
5) Ok, now, lets say I want to introduce an offset of 3 degrees lower. I press the down arrow keys which adjust the so-called offset reading --- and it goes from 20.0 to 17.0.
6) So, at this point, the left display shows 20.0 (current ambient) and the right display shows 17.0, which represents a -3 degree offset (not a 17 degree offset).
7) Now I press "Enter" and I plug in all the other thermocouples (4 in total).
All four displays now read approximately 17 degrees (not 20 degrees) as a result of introducing effectively a -3 degree offset.
9) That would all be ok...but now there is no way to get back to where you started. You've lost the reference to the current actual temperature....because the "offset"display isn't really just an offset from zero.
10) So, I try to get back to square 1 --- and I enter the "offset" mode again. But the left display doesn't show the ambient, it still shows 17 degrees -- and the right display shows 17 degrees. So how do I know what the current ambient is anymore? Yes, in this case I may remember that it was 3 degrees adjusted. But the temperature has changed in the meantime....and there seems to be no way to get back to square 1 where the input simply reads the current temperature without any arbitrary offset applied.
Hopefully this explains it better. It seems right now the only way to fix this is to adjust the offset to match the current ambient temperature as measured by another (separate) accurate temperature sensor (like my other Fluke meter)..
It really seems like they got the definition of offset wrong. To me, an offset should start at zero (meaning no offset is being applied). If I adjust it up 3, then I know I'm +3 above whatever the real temp is. But that's not how this seems to work. The offset is really a new temp reading. Very strange. I'm willing to accept there is something I'm doing wrong -- but the manual doesn't say how to eliminate the offset per-se and I can't seem to do it unless I have an absolute reference somewhere else. It seems as a result that I've lost the full calibration of this meter because the offset is something arbitrary that I was just playing with --- and I don't remember exactly where the ambient was at the time -- I'm probably back to within a couple of degrees -- but that's silly. Hopefully this additional detail helps if anyone has experience with this particular meter... thanks again!