| tautech:
Get the right tool for the job. |
| wmundstock:
--- Quote from: tautech on March 10, 2024, 10:02:54 pm ---Get the right tool for the job. --- End quote --- Thank you for the reply tautech, I do have other meters that are capable of doing this test, however, because I recently bought a 34401a, I wanted to test and learn its specifics ;D. |
| TimFox:
As you found in the manual, the programming for the 34401A will recognize a diode if 0.3 V < V < 0.8 V. This allows the machine to distinguish between a short circuit with 0.3 V > V a diode with 0.3 V < V < 0.8 V and an open circuit with 0.8 V < V The physical voltage with V > 0.8 V at 1 mA will be limited by the design of the current source, but is roughly 7 V max. These three possibilities are a reasonable set of conditions for a good or damaged silicon PN diode. |
| Kean:
--- Quote from: wmundstock on March 10, 2024, 09:53:40 pm ---In fact, I just referred to the manual one more time, and it is actually stated but I did not know how to read it. To Check Diodes Test current source: 1 mA Maximum resolution: 100 μV (range is fixed at 1 Vdc) Beeper threshold: 0.3 volts ≤ Vmeasured ≤ 0.8 volts (not adjustable) --- End quote --- I adjusted your bold highlighting. The text now in bold is the key thing. In diode mode, the measurement range is fixed at 1VDC - so this meter cannot measure Vf beyond 1V. |
| wmundstock:
Thanks for the replies, that clears out my questions. |