By the way, our supplier says that this bug is on all new MDO 2000/3000/4000. They are the same.
can you post a few pictures of the problem
OK,I post some pictures.
Test 1
This is a buck IC's LX pin waveform. Vin=5V, Vout=1V, Iout=0A. CH1 is TPP0250, this probe is compensated , CH2 is P9139B You could easy to see when fine scale is 1V, They are all the same. But when fine scale is 100mV, TPP0250 waveform is distort.
Test 2
This is a buck IC's LX pin wavefrom. At first Vin=14V, CH1 is TPP0250, The result of 5V,1V and 500mV fine scale is the some. And then, Vin=20V, CH1 is TPP0250. You could easy to see that in 500mV fine scale, the bottom of the wavefrom is no on zero. This is a false result. Because in other fine scale, they are on zero.
When you use DPO3000 and set the fine scale form 1V to 500mV?use 10x probe) , you can hear a tick of a relay, I think the on this time, scope change the amplifier. But in MDO2000/3000/4000, no this hardware. So it make Test 2 bug
If you are TEK's engineer, could you answer this problem?