Author Topic: Cheap 100:1 2KV Oscilloscope probes - any experience ?  (Read 7206 times)
Offline MiyukiTopic starter
  • Country: cz
    • Me on youtube
« on: March 20, 2017, 11:10:23 am »
Hello
I want to ask if you have any experience with cheap high voltage probes from ebay. Will they really work to 1200-1500V ?
I need them for smps diagnosis, yesterday I kill some 600V ones, yes they really die at higher voltage  :palm:  (there was some 1100V peaks and it hold it for short time then to spark gap)
Offline amspire
Re: Cheap 100:1 2KV Oscilloscope probes - any experience ?
« Reply #1 on: March 20, 2017, 11:36:57 am »
I have some Hantek T3100 100:1 100MHz 2KV probes that were pretty cheap, but you have to look carefully at the voltage derating. At 1500V, the maximum frequency is 2MHz. At 20MHz to 100MHz, the maximum voltage is about 400V.
If you are talking about a waveform with a lot of transients or noise, there may be a significant high frequency component. You can get the Hantek probes for about US$20.
Online Andy Watson
« Reply #2 on: March 20, 2017, 11:39:08 am »
I have an Isoprobe II  100:1 which I have used with 2.4kV signals - and I'm very happy with it. Sold (a year or two ago) under the name of Multi-Contact.
http://ec.staubli.com/AcroFiles/MA/SZ_RZ043_de-en-fr.pdf
« Last Edit: March 20, 2017, 11:40:45 am by Andy Watson »
Offline HighVoltage
« Reply #3 on: March 20, 2017, 02:41:57 pm »
I am using the TestTec series HV
Bandwidth:                    300MHz
Input Voltage Max:         2.5kV
Test Probe Attenuation:  100:1
For simple applications, this probe works very well.
Offline MrW0lf
  • Country: ee
    • lab!fyi
« Reply #4 on: March 20, 2017, 06:11:02 pm »
Have 2x Testec TT-HV 250 100:1
Attenuation Ratio: x100
Input Impedance: 100M?, 4 pF
Bandwidth: 300 MHz
Rise Time: 1,2 ns
Cable Length: 1,2 m
Compensation Range: 10..50 pF
Max. Input Voltage: 2500Vp CAT I, DC incl. Peak AC (derating
with frequency)
Scopes unburned so far.
Edit: Note, I used these only with low-energy HV as per CAT I spec!
"Measurement Category I:
This category is for measurements of voltages from specially protected secondary circuits. Such voltage measurements include signal levels, special equipment, limited-energy parts of equipment, circuits powered by regulated low-voltage sources, and electronics."
« Last Edit: March 21, 2017, 10:35:23 am by MrW0lf »
Offline Someone
    • send complaints here
« Reply #5 on: March 20, 2017, 10:17:06 pm »
And the associated derating of the voltage! Even these cheaper products are very clear about their limitations:
http://www.hantek.com/Manual/T3100_Probe_Manual.pdf
Above just 10MHz its only a 500V probe.
Offline joeqsmith
« Reply #6 on: March 20, 2017, 11:03:54 pm »
I tried some cheapo ones.  Video below.  I ended up making my own.  The ones I made I trust.
https://youtu.be/v1jJ1Z_fgFU?t=184
Home made ones are here
https://youtu.be/z7rnifjWd_U?t=531
« Last Edit: March 21, 2017, 02:37:50 am by joeqsmith »
Offline alm
« Reply #7 on: March 20, 2017, 11:48:07 pm »
Not sure if a circuit with high voltage transients that have already killed a probe is the place to use a cheap probe of dubious safety. Note that a fault in the probe could easily kill a scope or injure a person.
I would trust Testec more than random Chinese eBay probes. Some of the old Hameg probes looked very similar to the Testec ones, so at least at some point they made probes for reputable manufacturers.
Are we talking about the primary side or the secondary side of a high(ish) voltage power supply? Because if it is primary, you should probably use a probe rated for CAT II at the appropriate voltage.
The following users thanked this post: egonotto, jjoonathan
Offline joeqsmith
« Reply #8 on: March 21, 2017, 02:41:00 am »
Kill a scope for sure.  That's why I had the spark gap, the buffer amp and then another high speed clamp when I ran that cheapo probe.
  • Country: cz
    • Me on youtube
« Reply #9 on: March 21, 2017, 09:13:19 am »
I use cheap probes with cheap scope so if something fail no big deal (can be also easily repaired), in worst case I have old analog scope, it have valve input buffer with insane input voltage ratings
But this spark gap wit buffer amp and clamp is good idea
We are talking about primary side, in this this case it was PFC with 410V output and aux supply powered from this voltage, flyback topology with 1000V clamp, it operates at low frequencies 40-60 kHz
Online David Hess
  • DavidH
« Reply #10 on: March 22, 2017, 08:16:20 pm »
They work fine however be aware that they cannot be used safely at high voltages with AC coupling.
If a x100 probe has a 100 megohm input resistance which is all of the common inexpensive ones, then it has 99 megohms in series with the oscilloscope's input resistance of 1 megohms.  But when the oscilloscope's AC coupling capacitor is included, the capacitor will charge to the voltage at the probe's tip and the input voltage rating of the oscilloscope will be exceeded by several times causing high voltage breakdown in the oscilloscope possibly damaging it.
Better x100 probes have a shunt resistor preventing this problem and can be identified by a lower input resistance specification.
The following users thanked this post: egonotto, Miyuki, MrW0lf
  • Country: cz
    • Me on youtube
« Reply #11 on: March 23, 2017, 09:06:55 am »
Oh that is important info thanks  :-+
So I ordered two Hantek T3100
Offline ziakansi
« Reply #12 on: March 11, 2024, 11:47:31 am »
Hello Folks
I found this discussion quite useful. I am using T3100 Probes. Since, it is 100:1, so it should attenuate the voltage by the factor of 100. However, in my case it does not perform attenuation and gives me output as 1:1. I ordered four probes and all has the same output. Am I missing something here?
Please guide me how can we get the correct attenuation.
Online wasedadoc
« Reply #13 on: March 11, 2024, 01:05:34 pm »
Hello Folks
Sounds like you are missing the required 1 MOhm "load" on the probe output.  What are you connecting the probe to?
Online G-son
« Reply #14 on: March 11, 2024, 05:42:15 pm »
...you have to look carefully at the voltage derating. At 1500V, the maximum frequency is 2MHz. At 20MHz to 100MHz, the maximum voltage is about 400V.
I read somewhere that this derating is because the current flow through the capacitance of the probe cause a significant power loss = heat in the probe at high frequency. Is that the only issue?
I'm thinking if it's only heat we're worried about we should be able to more or less ignore the derating if we only get short high frequency, high voltage transients, with plenty of cooldown time between - but things tend to be more complicated than they seem so it might not be that simple.