Products > Test Equipment

Cheap Chinese Oscilloscope Preamp Tear Down/Review – Is It Any Good?

(1/4) > >>

gby:
I have had a need for scoping low level signals and been looking for a solution.  I have gone over the following EEVBlog threads:

   https://www.eevblog.com/forum/testgear/preamplifier-for-noise-measurements/
   https://www.eevblog.com/forum/testgear/quick-review-and-teardown-of-alphalab-lna10-low-noise-amplifier/

Unfortunately these choices are either fairly expensive, even used Ebay, or seemed to no longer be available.

I recently ran across an inexpensive Chinese unbranded “oscilloscope preamp” that I thought I would give a try.  Seems to be widely listed on Aliexpress, Ebay, etc for $60 on up deliverd.  In all cases the listing title was:

    10M Bandwidth Oscilloscope Differential Probe Signal Amplifier

I initially ordered from two different Aliexpress dealers and in both cases after listing the item as shipped a while later they asked me to cancel the order.  So, I am guessing that cheap Aliexpress shipping has trouble with the built in Lithium battery.  I finally got it from Banggood at:

   https://www.banggood.com/10M-Bandwidth-Oscilloscope-Differential-Probe-Signal-Amplifier-for-Weak-Electrical-Signal-Measurement-with-Metal-Shell-p-1959444.html?cur_warehouse=CN&rmmds=search

This post is a combined
  -   Tear down
  -   Review
  -   Request for help to finish reverse engineering and fixing

The published specifications of this T100 preamplifier are:

   Single-ended input voltage: -10V to +10V
   Differential Input Voltage: -10V to +10V
   Single-ended input impedance: 2M
   Differential input impedance: 4M
   Output Voltage: -10V to +10V
   Magnification: 1x/10x/100x
   Bandwidth:
      G=1 -- 10M
      G=10 -- 10M
      G=100 -- 2M
   Common mode rejection ratio: >80DB
   Noise : 5nV/√ @1KHz
   Power supply: built-in rechargeable lithium battery
   Charging interface: USB TYPEC
   Battery life: about 5 hours
   Working temperature: 0°C-40°C
   Working humidity: ≤90%RH
   Power consumption: ≤0.2W

TEAR DOWN:

The attached picture shows the item along with tear down front and back of the single internal PC board.
Assembled
Top side
Bottom side
Schematic
In reverse engineering it I have most of the circuitry traced out.  There are some mysteries/guesses because some the IC’s have the markings sanded off.  The high points are:
1.   Analog section is a single instrumentation amplifier with input direct from the input BNC’s and output direct to output BNC.
2.   Power supply is lithium battery to
     a.  Dc-dc boost to make +14.6Vdc
     b.  Dc-dc inverter to make -14.6Vdc
3.   Linear regulators to make +/-12Vdc
Lastly, see the attached partially finished schematic for this device.


REVIEW:

The Good:
  -   Compact
  -   Runs from internal rechargeable battery so easier to avoid noise from ground loops
  -   10 MHz BW for 1x and 10x with 2 MHz for 100x
  -   Reasonably low noise within caveat listed below

The Bad: (missing functionality)
  -   No ac coupling option
  -   No adjustable dc offset
  -   No adjustable low pass filter
  -   No adjustable high pass filter

The Ugly:
  -   -14.6 Vdc switching supply couples 70 mVp-p noise on the output which is way to big.  It must be fixed.
  -   Input pins direct to input BNC with no limiting impedance so more than 12V and device dies.
  -   Input impedance is set by 2 Meg resistor….would be better at 1 Meg so could use standard 10x scope probe.  Can use 10X probe and get known 5.5x atten at dc but standard probes don’t have enough ac compensation range for this pre-amp’s input capacitance.  Changing to 1 Meg will fix that.
  -   1x and 10x step response have HUGE overshoot.  Probably from stray capacitance on the gain adjust pins.
      o   1x has Po = 1.26 for 0.5 Vpp step or 252%!
      o   10x has Po = 0.70 for 0.5 Vpp output step or 140%!
      o   100x has no overshoot
  -   Output directly drives the output BNC and with a long cable the step response overshoot gets worse.  Included 0.5m cable adds 10% more overshoot.


NEXT STEPS:

1)   It would be great if anyone recognizes the instrumentation amplifier so that we could look at the data sheet.  It is:
  -   8 pin industry standard pinout instrumentation amplifier
  -   In SOIC package
  -   Gain equation is G = (1 + Rg/9.9k)
  -   Has 10 MHz BW at 10x
I can find many potential matches with the gain equation fixed/internal gain resistor at 24.7k or 3.0k.  But I have not been able to find one with 9.9k.

2)  The -Vdc switcher noise needs to be cleared from coupling to the output.  It does not get to the linear supply voltages so somehow this signal is direct coupling input the analog signals.  I highly suspect it is layout coupling.  Also, these noise spikes don’t change amplitude with changes in the gain so my suspicion is the VREF pin on the instrumentation amplifier that is tied to ground plane but near the -14.6V dc-dc supply output.

3)  Change the 2Meg input load R’s to 1 Meg so that 10x probes can work.  Probably just solder another 1 Meg R right on the bottom side at the input BNC’s.

4)  Is there a fix for the huge step response overshoot on 1x, 10x?  Probably have to wait until we find the data sheet for the instrumentation amplifier.

5)  Add a switch to get input ac coupling or maybe three-way switch and have DC, 1 Hz, 500 Hz with the last position used when there ac line noise you need to eliminate.  Any suggestions for a tiny three-way two pole switch that might work?

6)  Adjustable BW limit.  At least lower to 1 MHz and 100 kHz.  Easiest way to do this is on the output but it really wants to be done on the input which is hard to do and maintain high CMRR without insanely matched components and placement/layout.

7)  Any other thoughts from the EEVBlog community?

thm_w:
Good info.
There is a BG review:

--- Quote ---Very basic device, not a real lab instrument. Actually non-inverting input is the inverting one. 🤭 Metal shielding case is not grounded at all. On 100x range is a hughe phase shift and not compensated 50mV DC output offset. No frequency compesation at all, just basic opamp. +- 14.5V switching supply rails for opamp generates a lot of noise, no linear suppl. High figures of output noise in usable range, which is already very narrow. No real manual or measurement graphs available. Disapointed.
--- End quote ---

Here an alix link: https://www.aliexpress.com/item/3256804188345764.html
For the BG price, I think you could build a better device. Of course its going to take a lot of effort to design.

Why do you feel low and high pass are needed in hardware?

Someone:

--- Quote from: gby on September 02, 2022, 08:36:42 pm ---7) Any other thoughts from the EEVBlog community?
--- End quote ---
Don't bother wasting time on it, way too many problems to reverse engineer debug and then fix.

gby:
Low pass and high pass in HW is needed if interfering signals are large enough to cause clipping.  Once you clip you can't get the signal back no matter what post processing you do in the digital scope.  When using 1x probe/connection it would not take much ac line frequency leakage into the signal to cause saturation if the scope is set below 10 mV/div and you want to see below 100 uv/div.

thm_w:

--- Quote from: gby on September 02, 2022, 10:58:12 pm ---Low pass and high pass in HW is needed if interfering signals are large enough to cause clipping.  Once you clip you can't get the signal back no matter what post processing you do in the digital scope.  When using 1x probe/connection it would not take much ac line frequency leakage into the signal to cause saturation if the scope is set below 10 mV/div and you want to see below 100 uv/div.

--- End quote ---

How does Alphalab get away with 1000x gain when there is a low pass filter only?
I didn't see a 60Hz notch in there.

Navigation

[0] Message Index

[#] Next page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod