Products => Test Equipment => Topic started by: digitalrubbish on February 24, 2016, 12:56:11 pm
-
Dear friends
I acquired a CMU200 for my HAM radio lab. The unit however has an issue. The generator works fine but the spectrum analyser shows about 16 dB too low from 0 till 1200 MHz and above 2200 MHz. Between 1200 and 2200 all is fine.
Can anybody give me a hint what component could cause this. I am not afraid of digging into the boards as R&S repair service is way beyond my budget.
Thanks for all help and schematics of the input boards or mixer boards where obviously this fault is located.
Pedro Wyns ON7WP-AA9HX
direct mail to pedro.wyns@ the famous gmail dot com
-
Hi,
Just send you the service manual. A few questions:
- do you have the same results on both inputs?
- did you do a factory reset off all settings to be sure no correction table is loaded for the inputs (you can see this under "Connect Control" => "Ext. Att. Input" => "RX Connection Table"
-
I have this on all 3 inputs.
no correction table