Hi!
I've had a chance to study the circuit-diagrams now!
As Member CNCJerry said, there are a number of areas where you must be specially careful, both for your own personal safety and to prevent damage to sensitive semiconductor devices in the pre-amplifiers!
In particular this instrument requires a 550V H.T. supply in it's preamp circuit in order to provide a sufficently high source voltage for the Giga-ohm ranges, obtained from two voltage-doubler circuits in the power-supply fed from a 110-0-110V secondary on T101, providing +275V and -275V with respect to chassis-earth 0V.
First of all, you MUST be happy measuring high voltages and have a good-quality multi-meter with sound test-leads to make the tests suggested below!
1) Measure the voltage across C202/R211 with the multi-meter black lead on chassis/earth or the metalwork. The reading should be +275V DC +/-20V. If very low or missing, check D201/D202 for short or open-circuits and C201/C203 for open circuits. If the reading is particularly high you may have R211 open-circuit or a fault in the positive-side of the current-source in the preamplifiers, which I'll come onto later.
2) Measure the voltage across C204/R212 with the multi-meter black lead on chassis/earth or the metalwork. The reading should be -275V DC +/-20V. If very low or missing, check D203/D204 for short or open-circuits and C202/C204 for open circuits. If the reading is particularly high you may have R212 open-circuit or a fault in the negative-side of the current-source in the preamplifiers, which I'll come onto later.
3) Check the low-voltage supplies:-
a) +12V @ pin 10 of QA201 (a standard 723-type regulator) - any faults should lead to a test on the '723 itself (test by substitution), bridge rectifier D205 and it's reservoir capacitors C205 & C207 - you can use a standard 220uF to replace these.
b) -9V at TP2 - if low or missing check D206 for leaks or short-circuits and R201 (801 ohms - this odd value may be difficult to find but I think it should be available from a supplier of "E96" series resistor values);
c) +5V on TP4 - this is obtained from a TO-3 can type 5V monolithic regulator - if faulty (unlikely) an LM323K/STEEL (Mouser) can be used to replace it.
d) -12V at TP5 - any faults should lead to a check of QA202 and Q201, the rectifiers D209 & D210 and the reservoir electrolytic C213.
4) If all the power supplies test OK, the next test is to see of the meter can be made to read zero. Refer to the circuit diagram of the pre-amplifier stage drawing no. 25764E and unplug the Pre-amplifier Module PC361. Place a 100 ohm resistor between Pin 1 of J104 and Pin 4 of J104.
Switch the instrument on. The display should be within a few digits of zero - not more than +/-0.005. If you are still getting a "+1." overflow indication the fault may be in the positive/negative pre-amplifier source circuits Q105/Q108 (positive) or Q106/Q109 (negative), or a fault in the associated resistors R126-R136. If the transistors are leaky or test unserviceable, you may need to get a "parts mule" (another unit the same) as these are specially selected Keithley low-leakage devices.
***CAUTION*** - make sure the 275V smoothing capacitors C202 & C204 in the PSU are completely discharged (zero volts) before making any component tests in the pre-amplifiers!
The resistor values should be stocked by Mouser, etc.
5) If you can't find a problem with Q105/Q108 & Q106/Q109, leave the pre-amplifier PC361 out and transfer the 100 ohm test resistor to J118 & J119 and again switch the instrument on and test for a very small or zero reading. If you still get the "+1." overflow indication with a 100R across J118/J119 there is a fault further forward in the A/D Converter and digit-drive logic, which I'll cover in a further post if needed!
6) If you DO get a zero reading (or very close to it) when the 100R is connected between J104-1 and J104-4 the fault is either in the Preamplifier Module PC361 or all the associated range-switching prior to that, which, again, I'll cover in a later post if needed!
Let us know how you get on so far!
Chris Williams
PS!
The manual is too large to post as an attachment at about 4.4M but it is freely downloadable from k04BB or "Dxdevs" that specialise in Keithley DMM repair, and possibly Electrotanya as well! Print it out and have a good read of it to begin with - long experience has taught me that trying to fault-find gear off a computer or lappy screen is a frustrating exercise - an open book with it all laid out in front of you is so much easier!