Products > Test Equipment
Conversion of 500MHz TDS744A to 1GHz TDS784A
<< < (7/24) > >>
nctnico:
How about getting a simple mmic based amplifier?
Jay_Diddy_B:
Hi,

As well as needing high amplitude, the signal needs to be levelled. That is the amplitude needs to  be constant over the frequency range.

You may trouble if you use a MMIC amplifier.

Regards,

Jay_Diddy_B
nctnico:

--- Quote from: Jay_Diddy_B on July 01, 2016, 04:26:37 pm ---Hi,

As well as needing high amplitude, the signal needs to be levelled. That is the amplitude needs to  be constant over the frequency range.

You may trouble if you use a MMIC amplifier.

--- End quote ---
If you use an RF generator you can adjust the amplitude so the output of the MMIC is correct.
edavid:

--- Quote from: Jwalling on July 01, 2016, 02:38:23 pm ---Looks like I'll need to score an amplifier - I was looking at HP 8447D amps but they are a bit pricey!
EDIT: not only that, but the output is only +7dBm - never mind...

--- End quote ---

The 8447E is the power amp version, with +12.5dBm output.

Jwalling:

--- Quote from: edavid on July 01, 2016, 04:38:50 pm ---
--- Quote from: Jwalling on July 01, 2016, 02:38:23 pm ---Looks like I'll need to score an amplifier - I was looking at HP 8447D amps but they are a bit pricey!
EDIT: not only that, but the output is only +7dBm - never mind...

--- End quote ---

The 8447E is the power amp version, with +12.5dBm output.

--- End quote ---

And that's not enough either.  :(

Jay
Navigation
Message Index
Next page
Previous page
There was an error while thanking
Thanking...

Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod