| Products > Test Equipment |
| Conversion of 500MHz TDS744A to 1GHz TDS784A |
| (1/24) > >> |
| Jay_Diddy_B:
Hi Group, In this thread: https://www.eevblog.com/forum/testgear/enabling-option-1m-extended-acquisition-memory-tektronix-tds754a/msg565086/#msg565086 I documented the process for enabling the extended acquisition memory option on TDS7xx scopes. In this thread I am going to show how I successfully converted a 500 MHz TDS744A to the 1GHz TDS784A. Tools and Equipment needed for this conversion You need a PC fitted with a GPIB board to run the Tektronix Field Adjustment software. This is described here: https://www.eevblog.com/forum/testgear/tektronix-tds700a-field-adjust-software-help-needed/msg563939/#msg563939 You need an accurate voltage source. Tektronix use a Data Precision (Analogic) 8200. You need +/- 9.5V, +/-0.95V and +/- 0.095V. You need a signal generator to 1.005 GHz. Tektronix uses the SG504 with the levelling head. I substituted an Agilent E4425B Base line measurements I used my Transmission line avalanche pulse generator to measure the risetime of the unmodified scope: The pulse generator is documented here: https://www.eevblog.com/forum/projects/transmission-line-avalanche-pulse-generator/msg185639/#msg185639 The measured risetime was 650ps this is consistent with a 500 MHz scope. Configuration Resistors Tinhead documents the configuration resistors on the back of the Acquisition board in this message: https://www.eevblog.com/forum/testgear/tek-tds-3054b-vs-tds-744a/msg539818/#msg539818 So I removed the Acquisition board. Here is a picture of the board configured for a TDS744A: And after I removed R1064, the board is configured for a TDS784A: I put the scope back together and the boot screen changed to this: Excited by this I measured the risetime again, there was no change: Input Hybrids I started researching eBay to see if the TDS784A used different ceramic hybrids in the input attenuator. I found some listings for the H2462G (pulled from a TDS744A) and H2462J (pulled from TDS784C). I was thinking that may the input hybrids were limiting the frequency response. These are the photographs from eBay: They looked very similar to me. Somehow it didn't make sense to use the same part number with a revision code for different bandwidths To be continued ... Regards, Jay_Diddy_B |
| Jay_Diddy_B:
Continuation.. Not really satisfied that the input hybrids were causing the bandwidth limitation I decided to look closer at the Acquisition Board. I found this picture of a TDS784A Acquisition Board on eBay: If you look really closely you will see that C1266, C1267, C1268 and C1269 are not populated. I looked at the schematics in the TDS520B Component Level Service manual to see what these capacitors are used for: In the second schematic these capacitors are connect across the differential signal leads. I suspected that these capacitor are there to reduce the bandwidth. I had a look at the Acquisition Board in my TDS744A and they were fitted: So I carefully removed these capacitors and saved them so I could put them back. After removing the capacitors you have to run the Signal Path Compensation (SPC). I checked the risetime again and was significantly faster. To be continued... Regards, Jay_Diddy_B |
| Jay_Diddy_B:
Continuation ... Having modified the hardware I used the Tektronix Field Adjustment software to calibrate the scope. The software calls for the use of a SG504 levelled signal generator. I don't have one of these. I used an Agilent E4425B RF Generator and very high quality cable. I found I had to increase the generator output by 5% when the software called for frequencies higher than 500 MHz to compensate for losses in the cable. After a couple of false starts I was able to calibrate the scope. Risetime after Modification and Calibration Here is a picture of the risetime: The risetime is 360ps, this is consistent with a 1 GHz scope. Sine wave measurements Here are some sinewave measurements confirming the increased bandwidth: I started with a 6 MHz signal and adjusted the output of the signal generator for a reading of 100mV RMS on the scope. I then increased the frequency to 100 MHz and the scope read 100mV At 500 Mhz still reading 100mV At 760 MHz the reading dropped to 86mV (-1.3 dB) At 1 GHz the reading dropped further to 76.6 mV (-2.3dB) It looks like a successful conversion :D :D :D Enjoy !!! Jay_Diddy_B |
| dxl:
Hi, i can confirm the modification. I have a TDS754D, which i upgraded to TDS784D by changing the ID resistors. However, the calibration failed, because freqency response was out of spec (1 GHz was low by ~-4 to -5dB). After removing the capacitors, the calibration runs now flawlessly, and the Field Adjustment software says it passed calibration. :D Regards |
| nctnico:
Great hack! I feel so stupid selling my TDS744A earlier this year |O |O |
| Navigation |
| Message Index |
| Next page |