@Tmax
I know exactly what you are talking about - there is absolutely no low pass filtering on the GPP series power supplies. I have both the GPP-4323 and the GPP-3323 and they act identical. As you state, unless you are connecting a very static or resistive load the current monitoring is pretty much useless. For example, the other day I connected a 24V fan to the power supply just to see what current it draws. Just about any DMM I put inline would average the current and give me a pretty good measurement of the current. But the Instek supplies just bounce all over the place. Or if you are trying to find the current consumed by a digital board that has fluctations in the current the power supply is useless.
I've been tempted to send an email to GWInstek to see if they would consider a FW update for this, but I doubt they would do anything.
There are 2 ways to do this - one way is hardware by putting a cap on the resistor divider that goes to the A/D and the other would be software. I suspect the hardware method will be difficult without a schematic. The firmware method would be simple, just average a number of samples. For example, you could take 10 samples and average them. It would actually be fairly easy to do 8 samples, sum them, and then shift the accumulator 3 bits to the right. But Instek would have to be willing to do this.
Best,