Hi
I just bought on Ebay an HP 8594e SA (my first SA) , the unit is in excellent exterior condition and it's stated in working condition, but i have some problem with the ampl calibration test and i hope someone here will be enough familiar with this unit to be able to help me.
After the unit is on for about an hour i did the confidence test and the frequency cal test and i got no error but when the Ampl cal test run it's stop at the 1Khz BW and say CAL RES BW AMPL FAIL SRQ 110. I checked in the trouble shooting manual and this can be 3 thing, Filter need calibration, a board is fry, or it's just a lack of calibration correction in the memory. In the trouble shooting manual this is the text about that
CAL: RES BW AMPL FAIL (H)
During the CAL AMPTD routine, the insertion loss of a resolution
bandwidth filter, relative to the amplitude of the 3 kHz resolution
bandwidth filter, was greater than 2.2 dB.
Related Assemblies
A7 analog interface, A11 bandwidth filter, A13 bandwidth filter, A14 log
amplifier/detector.
Troubleshooting Hints
The CAL AMPTD routine will stop at the resolution bandwidth that
failed. If the A11/A13 bandwidth filter assemblies require realignment,
use the crystal and LC bandwidth filter adjustment procedures in
Chapter 2.
If a defective bandwidth filter board is suspected, refer to “IF
Power-Level Measurement” in Chapter 5.
i found in the Agilent forum this discussion
http://www.home.agilent.com/owc_discussions/thread.jspa?messageID=59931&tstart=0 but it seem to be more related to the option 130 (which i have but can't test that since it stuck at 1khz).Now if anyone can tell me if there a easy and not intrusive test or thing i can do to be sure it's only a calibration issue, i still have some time to return the unit but i don't want to pay someone half the price i paid to troubleshoot this unit to found it was a hardware failure.
When i test manualy i can see a drop in the 1 khz and 300 hz range but this disappear if i enable the correction option in the cal menu. So i'm stuck here, calibration or hardware error?? since i still get an error for the AMPL cal test.
Thanks in advance for your help.