Thanks you for the video!
I have some questions and suggestions:
1. By measuring the resistances, everything is fine, only you did not show the closure of the test contacts to each other and the stability of the zero readings on the display? I would also like to see the calibration process?
2. I showed on the photo 1 place where the unit of measurement of reactance is indicated not completely "m", "G". This is clear to me, but may not be very clear to another user. I understand that there is not enough space on the display to accommodate all the information, so I suggest slightly reducing the size of the beep icon. You can simply display the speaker icon when the sound is on and the strikethrough icon when the sound is disabled.
3. I showed in photo 2 the moment that when measuring a capacitance of less than 1 μF, it is desirable to output not Rs, but an important secondary parameter D. The Rs parameter is usually shown when measuring a capacitance of more than 1 μF.
4. I showed in photo 3 that when measuring the inductance, they show the secondary parameter Q if the Q-factor of the coil is more than 1 and show the parameter Rs if the Q-factor of the coil is <1. I am also sure that the test frequency of 10 kHz will not be enough for stable measurement of small values of inductance. This will require a frequency of 50-100 kHz.
This is just my opinion, I do not insist on 100% certainty.
[/quote]
Thanks a lot for your suggestion!
1. Cal process support selfcal/open/short calibration, as the first attached
2. Thanks a lot for your suggestion, we will try to modify that
3/4. D/Q mode is supported, which need go to measure menu and select R mode or Q/D mode, as 2 and 3 attached, 50k-100kHz is our next goal, which needs more support, we focus on the 10kHz on this version