« Reply #650 on: April 11, 2024, 09:10:45 pm »
I was a little away from Eevblog because of some personal problems.

Nice work!

Thank you very much  :D

Well I went into this thread once again, this time because I was having problems updating my ST42 so it's funny that you say "hope you are the only unlucky guy who has this problem" because I'd like to introduce myself: Hi, I'm another unlucky guy, it seems :/ I'm getting exactly the same error, and I'm not really happy to hear it's a hardware problem. Is replacing the chip the only solution?

EDIT2: I would like to add that I remember updating the firmware when I was having the screen glitches (which I also reported here). And I don't remember the process giving me errors that time. Maybe this is not necessarily a hardware problem (in my unit at least) but rather that the sofware gives the same error under different conditions (including faulty hardware).

EDIT3: Ok, "solved" the issue and I'd like to keep a record here despite being a very silly mistake from my part. Apparently the intimidating error shown by the updater is the same in case of any problem communicating with the device. In my case it was either a driver issue or a cable issue, because changing the cable and usb port I got it to work.
BUT I also got the same error during one attempt in which I didn't enter into bootloader mode correctly. The COM port appeared on my computer but it wasn't in *actual* bootloader mode.

So I'd like to leave this post here as a warning for others: Even if you get the same error, it does not mean it's a hardware error. As obvious as it is now, the firmware updater will give the same error in any case in which the device is not responding back.
I'm happy to read that, as it was finally only a software/setting problem  ;)
For me, the RX pin of the chip was shorted to ground  :palm: Since I put the new one, no problem at all  :-+

For @shannon: I have checked the new diode signal: it is now super clean  :-+
I have not used it a lot since my fix, so I don't know what is new in the 1.54 version ?
@Shannon: Do you have a changelog or something like that ? With that, I can continue to suggest some ideas I could have  :)


 

Offline ceut

  • Regular Contributor
  • *
  • Posts: 132
  • Country: fr
Re: Design a New Precision LCR Tweezers with Much Lower Cost(Video Attached)
« Reply #651 on: April 11, 2024, 09:19:02 pm »
@Shannon how difficult do you think it would be to implement a special electrolytic capacitor testing mode? Electrolytic caps are polarized, and thus the correct way to test them is (I think) to bias the test signal to the Vpp/2 level, so that it always stays above zero instead of the standard AC signal of +/- Vpp/2.

I'm not sure what difference it will make though, but it just seems to me that applying a reverse polarity voltage to electrolytic caps, even if low and only for a half period, isn't quite right -- might be from the cap damage or the measurement precision perspective, or both. There are some LCR meters that implement this biased signal mode, e.g. East Tester ET432/433. I'm not able to find any good info on this topic, only vague suggestions that it's "better" to test the electrolytics with a positively biased signal, and it gives better precision.

It would be great to have that.
Also, as I was looking for a LCR Tweezer before Tautech told me about Shannon's one, here is the LCR-MPA function table, which I think may be added for some of them (the blue one would be great  8) ) ?
 

Online KungFuJosh

  • Super Contributor
  • ***
  • Posts: 1670
  • Country: us
  • TEAS is real.
Re: Design a New Precision LCR Tweezers with Much Lower Cost(Video Attached)
« Reply #652 on: April 21, 2024, 12:35:36 am »
No need to wait any longer, just update your tweezers right now. >:D

Is there a new firmware release? Your website still says 2023/5.

Just upload the new firmware a momenta ago, the website information has not been update.
And there are some other modification is under processing. ;)

Hey Shannon, any update regarding the newer firmware? The OTA tool still downloads the older firmware, and the website also says 5/23.

Also, it's probably a good idea to add Shannon ST42 Tweezers to this thread title by now. 😉
"I installed a skylight in my apartment yesterday... The people who live above me are furious." - Steven Wright
 
The following users thanked this post: Kean, Shannon

Offline kladit

  • Contributor
  • Posts: 49
  • Country: de
Re: Design a New Precision LCR Tweezers with Much Lower Cost(Video Attached)
« Reply #653 on: April 23, 2024, 10:40:11 am »
I managed the update on my linux now.

My tweezer is detected by the kernel (type dmesg to see) as:

usb 1-4: Product: USB Serial
ch341 1-4:1.0: ch341-uart converter detected

lsusb shows it as:
ID 1a86:7523 QinHeng Electronics CH340 serial converter

udev mapped it to (type dmesg to see) :
usb 1-4: ch341-uart converter now attached to ttyUSB0

Now take the shutted down tweezer, press its navigator button upwards and put in the usb-connector
of the usb-cable already conected to your computer.

Now cd to the dir where you have the unzipped  fw_update-linux programm and type :

mkdir ./tmp; export TMPDIR=./tmp; ./fw_update-linux  ttyUSB0

Now the firmware gets loaded over the internet and is flashed to the tweezer.


« Last Edit: April 27, 2024, 12:01:51 pm by kladit »
 
The following users thanked this post: Shannon

Offline m.abdelwanis

  • Contributor
  • Posts: 12
  • Country: eg
Re: Design a New Precision LCR Tweezers with Much Lower Cost(Video Attached)
« Reply #654 on: April 30, 2024, 04:42:28 am »
Any info about ESR table or calibration ?
 

Offline ShannonTopic starter

  • Regular Contributor
  • *
  • Posts: 234
  • Country: hk
Any info about ESR table or calibration ?

Could you give us more details about your idea?

ST42 has the independent ESR measurement mode, so what does the ESR table mean?

Does it mean if ST42 show a ESR measurement table after testing many DUT?

And ST42 has four calibration mode: "self-cal" "short-cal" "open-cal" "factory", for the calibration process, please refer to
Nature and Nature's laws lay hid in night:
God said, Let Newton be! and all was light.
 

Offline ShannonTopic starter

  • Regular Contributor
  • *
  • Posts: 234
  • Country: hk

Hey Shannon, any update regarding the newer firmware? The OTA tool still downloads the older firmware, and the website also says 5/23.

Also, it's probably a good idea to add Shannon ST42 Tweezers to this thread title by now. 😉
We will put the release information on the website.

And thanks a lot for your advice, the title has been updated.
Once again, it proves that our business sense is far away from good enough :palm:
Nature and Nature's laws lay hid in night:
God said, Let Newton be! and all was light.
 
The following users thanked this post: KungFuJosh

Offline ShannonTopic starter

  • Regular Contributor
  • *
  • Posts: 234
  • Country: hk
@Shannon how difficult do you think it would be to implement a special electrolytic capacitor testing mode? Electrolytic caps are polarized, and thus the correct way to test them is (I think) to bias the test signal to the Vpp/2 level, so that it always stays above zero instead of the standard AC signal of +/- Vpp/2.

I'm not sure what difference it will make though, but it just seems to me that applying a reverse polarity voltage to electrolytic caps, even if low and only for a half period, isn't quite right -- might be from the cap damage or the measurement precision perspective, or both. There are some LCR meters that implement this biased signal mode, e.g. East Tester ET432/433. I'm not able to find any good info on this topic, only vague suggestions that it's "better" to test the electrolytics with a positively biased signal, and it gives better precision.

It would be great to have that.
Also, as I was looking for a LCR Tweezer before Tautech told me about Shannon's one, here is the LCR-MPA function table, which I think may be added for some of them (the blue one would be great  8) ) ?
@ceut @shapirus Thanks for your information ;D

The effect of DC bias on capacitance is a standard spec for those ceramic capacitors, electrolytic capacitors, etc.
It may not be easy to implement because a table is usually used to describe it. Taking ceramic capacitors as an example,
The abscissa of a chart is the DC voltage value, and the ordinate is the loss of capacitance. Therefore, when we consider providing DC bias, we need to implement an adjustable and wide-range DC bias voltage, which is reasonable.

However, as a low-power high-precision device, ST42 is basically unlikely to provide a higher DC bias, because the operating voltage of the core device is 3.3V.

We also checked the parameters of the EST43 series and found that the internal bias is adjustable, but the adjustable range is 0~500mV. Such a bias range is also difficult to meet real test requirements.
The most common application in my mind, what is the capacitance of the capacitor under 12V/5V/3.3V/1.8V DC bias.

But you all know, when you know one cap value of two serial cap and the serial cap total value, you can calculate the other cap value.
This principle could help you measure the DUT cap under any DC bias voltage. This test method makes me feel exited >:D
« Last Edit: May 05, 2024, 03:35:42 pm by Shannon »
Nature and Nature's laws lay hid in night:
God said, Let Newton be! and all was light.
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1473
  • Country: ua
However, as a low-power high-precision device, ST42 is basically unlikely to provide a higher DC bias, because the operating voltage of the core device is 3.3V.

Well, my point was about implementing DC bias that's not high or adjustable (even though that would be nice to have), but only just so high as to shift the test signal above zero, so that it does not change polarity, which, as I assumed, might make sense for electrolytic caps.

Maximum Vpp swing of the test signal in the capacitance mode is only 1.0 V, so it's swinging from -500 mV to +500 mV. With a DC bias of +500 mV it would swing from 0 V to +1.0 V. That should still be well within the capabilities of the device operating at 3.3 V.

As it is now, electrolytic caps are tested with a signal that changes between forward and reverse polarity every half cycle. The level is probably too low to damage the capacitor under test, but my concern was about whether it's ok to use signal with alternating polarity to measure polarized caps (which by definition should never be reverse biased). Maybe it is, I don't know. But there's probably a reason why some LCR meters have dedicated modes for electrolytic caps.
« Last Edit: May 05, 2024, 05:23:33 pm by shapirus »
 

Offline ceut

  • Regular Contributor
  • *
  • Posts: 132
  • Country: fr

Could you give us more details about your idea?

ST42 has the independent ESR measurement mode, so what does the ESR table mean?

Does it mean if ST42 show a ESR measurement table after testing many DUT?


I think the idea is (again) like the MPA tweezer in the screenshot I have made here from their user manual.
I think it is the rejection table.

Also, I think (?) the ESR Table A and B could be used as reference for checking Cap. , with any LCR Meter which shows ESR value .

I don't know if it would be a great addition, because in MPA: we have to set the capacitor voltage manually to have this rejection value, and this is not very convenient IMO.
That's why I have not pointed it out on my previous message  :)

But all other ideas I have put in blue would great addition to the ST42  ;) :-+
« Last Edit: May 12, 2024, 09:40:51 am by ceut »
 

Offline ShannonTopic starter

  • Regular Contributor
  • *
  • Posts: 234
  • Country: hk
However, as a low-power high-precision device, ST42 is basically unlikely to provide a higher DC bias, because the operating voltage of the core device is 3.3V.

Well, my point was about implementing DC bias that's not high or adjustable (even though that would be nice to have), but only just so high as to shift the test signal above zero, so that it does not change polarity, which, as I assumed, might make sense for electrolytic caps.

Maximum Vpp swing of the test signal in the capacitance mode is only 1.0 V, so it's swinging from -500 mV to +500 mV. With a DC bias of +500 mV it would swing from 0 V to +1.0 V. That should still be well within the capabilities of the device operating at 3.3 V.

As it is now, electrolytic caps are tested with a signal that changes between forward and reverse polarity every half cycle. The level is probably too low to damage the capacitor under test, but my concern was about whether it's ok to use signal with alternating polarity to measure polarized caps (which by definition should never be reverse biased). Maybe it is, I don't know. But there's probably a reason why some LCR meters have dedicated modes for electrolytic caps.
I try to find some parameters for polar capacitors that require attention to short-term reverse voltage, but couldn't find any.

Agilent's E4980A LCR meter supports DC bias, and supports both voltage bias and current bias. This is its manual https://www.keysight.com/us/en/assets/7018-01355/data-sheets/5989-4435.pdf.
Voltage bias range +/-40V, current bias range +/-100mA.
But the terrifying 40A can be achieved through options, here is 4284A manual https://www.testequipmenthq.com/datasheets/Keysight-42841A-Datasheet.pdf.
Obviously these are testing the impedance changes of capacitors and inductors under the influence of DC.

In my understanding, polar capacitors (electrolytic capacitors, tantalum capacitors) can accept AC signals when DC is 0V.
I wonder if any netizens know more about whether a polar capacitor can accept AC coupling at 0Vdc. If it can, what is the frequency of this AC signal?
In the extreme case, when the AC signal frequency is extremely low, there is no difference from DC.
This is something I haven't figured out yet :palm:
Nature and Nature's laws lay hid in night:
God said, Let Newton be! and all was light.
 

Offline ShannonTopic starter

  • Regular Contributor
  • *
  • Posts: 234
  • Country: hk

Could you give us more details about your idea?

ST42 has the independent ESR measurement mode, so what does the ESR table mean?

Does it mean if ST42 show a ESR measurement table after testing many DUT?


I don't know if it would be a great addition, because in MPA: we have to set the capacitor voltage manually to have this rejection value, and this is not very convenient IMO.
That's why I have not pointed it out on my previous message  :)

But all other ideas I have put in blue would great addition to the ST42  ;) :-+

In ESR Table Usage.jpg, "The second number is the selected operating voltage of the capacitor inmenu under C voltage:6.3,10,16,20,30,50-63, 100, and 160+Volts."

This DC bias voltage is really very high.

Do you have the whole user manual?

BTW, another batch of Curved Tweezer Tips has been produced and we have added the option of the ST42+ curved Tweezer Tip Kit in the shannon tweezers store https://www.aliexpress.com/item/3256804053554452.html?spm=a2g0s.12269583.0.0.4cf7356cIFgu1J#nav-review
« Last Edit: May 16, 2024, 10:40:08 am by Shannon »
Nature and Nature's laws lay hid in night:
God said, Let Newton be! and all was light.
 

Offline ceut

  • Regular Contributor
  • *
  • Posts: 132
  • Country: fr

Do you have the whole user manual?

BTW, another batch of Curved Tweezer Tips has been produced and we have added the option of the ST42+ Elbow Tweezer Tip Kit in the shannon tweezers store

Hello Shannon  :D
Yes, I give you links of the usermanual: https://www.lcr-reader.com/LCRMPAmanual.pdf
and the datasheet: https://www.lcr-reader.com/lcrreadermpa_specsheet.pdf

This tweezer is in fact a clone of the original russian one, the "NV15/HB15" model (which has FW update):
http://www.rlc-esr.ru/attachments/article/64/NV15_v1.1.pdf
http://www.rlc-esr.ru/index.php/ru/pintset-nv15
The price is good, but they don't sell it outside of Russia (I have e-mailed them).
I have taken a lot of time to understand it, with the help of google translate, because I really don't understand a single russian word  :-DD
The canadian company have copied, and sold it at about 3x the price  :--
The russian guys have also made newer version of their product.

Hope that will help you to improve the famous ST42  :-+