Hi...
I'm trying to determine effective power consumption by integration, over a complete cycle at 50Hz,
and not over 1 second (as it is in the majority, if not all, oscilloscopes) .
The video "How & Why to use Integration on an Oscilloscope" from Dave
was very elucidate but there is something I'm still missing.
Scope screenshot is at:
photobucket.com/gallery/user/miguelgpbs/media/bWVkaWFJZDoxMzAzNTk3NTQ=/?ref=
anyway, the image is also attached here...
I'm using a Rigol DS1054z
* Yellow (CH1) : Mains sine wave voltage, 230Vac rms  50Hz
* Cyan (CH2) : Current, measured through a 1 ohm series resistor.
* Purple : Voltage (Yellow CH1) times Current (Cyan CH2) math function to compute power
* Red vertical lines are the imposed area of measurement
As the current is an irregular waveform, consequently the power is also a irregular one.
To integrate this irregular power waveform over a complete cycle, I used the area measurement function of this oscilloscope,
confined between the red lines (this gives me immediately the final integrated value) .
The result value for the integration (area calculation) was 152mWs (green bordered at bottom right side).
Question is:
The 152mWs, as the unit of measurement itself states (mWs), is valid for a time of 1 second.
So, as long as I understood, this result is not the effective power for a 50Hz frequency.
Given this, how to calculate the effective power consumption, from the result 152mWs, for the
power waveform (Purple) over a complete 20ms cycle (50Hz) ?
Thanks in advance,
Miguel Garcia
« Last Edit: October 28, 2018, 12:31:52 am by miggarc »
Offline PTR_1275
Is the scope automatically allowing for that figure to be 1second sample time?
What happens if you increase to 2 cycles, or 10, or a whole second. Does that figure change by the order of magnitude you expect?
Offline MrW0lf
  • Country: ee
    • lab!fyi
Nope, it's just equivalent amount of energy compared to instantaneous power of 152mW during 1 second. 1J = 1Ws. Think kWh. Does it matter in what amount of time you spent 1kWh? You could have used 100W device during 10h, or 1kW device during 1h. Bill is just the same.
You could switch on actual integral function trace on CH1*CH2 to cross check area calculator.
« Last Edit: October 28, 2018, 10:09:18 am by MrW0lf »
Offline miggarc
  • Country: pt
I will try to increase number of cycles for integration and then let you know.
Many Thanks for your answer
Offline miggarc
  • Country: pt
Hi
But... Dave Jones said on his video, at 11m 02s, we must not confuse this units because it is not an energy unit.
Could he be wrong?
https://youtu.be/Dh0xYu8YvaE?t=662
In fact, if I convert this result from energy to watt, it gives me an absolute erroneous value.
Try it at:
https://www.rapidtables.com/calc/electric/kwh-to-watt-calculator.html
Inputting those values:
kWh : 0.000152  (152mW=0.000152)
h : 0.000278  (1s = 0.000278h)
Result is 546.7 W
which is out of question because the circuit related to the scope measurements
has an absolute maximum power of 25W .
Thanks
« Last Edit: October 28, 2018, 10:52:20 am by miggarc »
Offline MrW0lf
  • Country: ee
    • lab!fyi
He is talking about Vs. You are measuring Ws.
If it is of any help here is some post about power measurements:
https://www.eevblog.com/forum/testgear/what-you-do-with-math-channels-on-oscilloscope/msg1855935/#msg1855935
But beware this talk is mostly around sinusoidals and terminology might be not fully correct since I'm not native english speaker, however you might find some general approach how to "take apart" such problems for better understanding.
0.152Ws/0.02s=7.6W
« Last Edit: October 28, 2018, 10:57:59 am by MrW0lf »
Offline miggarc
  • Country: pt
OK MrW0lf
I will take a look there.
Thanks again
Offline David Hess
  • DavidH
Note that integration of power produces a measurement of energy and *not* power hence milliwatt-seconds instead of just milliwatts.  1 watt-second is 1 joule of energy.
152 milliwatt-seconds is the energy over one powerline cycle.  Converting watt-seconds to watts means dividing by time or in this case, 1/50 seconds which is the same as multiplying by 50.  So 152 milliwatt-seconds divided by 1/50 seconds yields 7600 milliwatts or 7.6 watts.
Offline miggarc
  • Country: pt
Hi David Hess
Although I've already figured out how to obtain power over an entire cycle from the 152 mWs energy result,
many thanks for your answer...
Offline macboy
Hi
Could he be wrong?
Try it at:
Result is 546.7 W
Thanks
The calculator asks for kWh (kilo Watt hours) and you entered kWs (kilo Watt seconds), which resulting in the wrong answer.
To convert mWs to kWh, divide by 1e6 mW/kW (mW units cancel and kW remain), then divide by 3600 s/h (s units cancel, h remains).  So 152 mWs is in fact 4.22e-8 kWh or 0.0000000422 kWh.
Offline bson
Did you really run 230VAC mains through a 1Ω resistor?
Offline miggarc
  • Country: pt
Hi
Yes, that's the idea ...
Offline Brumby
52,900W?!!  That's the RMS power you would get for each quarter cycle.  That's going to need a seriously low impedance source as well as one helluva resistor or some very short ON times.
.... or is this a purely theoretical exercise?
Offline macboy
You have missed the fact that the 1 ohm resistor is the current sense shunt, not the load.
Offline miggarc
  • Country: pt
I forgot the measurement schematics.
Here it is.
PWRS is obviously a power supply, a switch mode one (SMPS) .
At is of interest here is all to the left of the PWRS, including itself:
This is the side of the circuit to witch all values are related to.
I'm measuring the PWRS efficiency.
« Last Edit: November 02, 2018, 03:55:53 pm by miggarc »
Offline Brumby
 :palm:
Excuse me, I have to go find something....  Ah here's one:  |O