Author Topic: Difference between Rohde & Schwarz RTM3k-COM4 and RTA4k-COM4?  (Read 2128 times)
Online egonotto
Hallo,
what are the difference between RTM3k-COM4 and RTA4k-COM4?
So far I only found 6 distinction.
1. The timebase accuracy:
For RTM3k-COM4:
After delivery/calibration, at +23 °C      ±2.5 ppm
During calibration interval                     ±3.5 ppm
For RTA4k-COM4:
After delivery/calibration, at +23 °C      ±0.5 ppm
During calibration interval                     ±1.0 ppm
2. The memory:
For RTM3k-COM4:
40 Msample/(logic) channel
200 Msample in segmented
memory mode/(logic) channel
For RTA4k-COM4:
100 Msample/(logic) channel
500 Msample in segmented memory
3. DC gain accuracy
4. Counter: 6 vs 7 Digit
5. The cost.
6. The name.
Best regards
egonotto
« Last Edit: May 22, 2019, 03:08:09 am by egonotto »
Online egonotto
Re: Difference between Rohde & Schwarz RTM3k-COM4 and RTA4k-COM4?
« Reply #2 on: May 22, 2019, 03:59:40 am »
Hallo,
thanks. I read the paper already and find the stated differences.
I want to know whether it exists more differences.
Best regards
egonotto
Offline maginnovision
« Reply #3 on: May 22, 2019, 04:12:21 am »
I thought the RTA had a lower noise floor but I can't remember where I found that. Considering the COM4 deal difference is only 2k, I think the listed items are probably about right.
Offline JxR
« Reply #4 on: May 22, 2019, 04:55:12 am »
Hallo,
Best regards
egonotto
I would just contact R&S and ask them.  There may be no one here that actually has a RTA4000.  One would have to be very familiar with both devices to know what undocumented differences there are between them, if such things exist.  There really is no reason to assume R&S is trying to hide the specifications of either device.  I would assume the memory specifications are the main difference, and reason for the different price points.
You could try reading the firmware notes also.  Its possible that there are firmware differences between them, depending on the update cycle.
Offline YetAnotherTechie
  • Country: pt
« Reply #5 on: May 22, 2019, 07:49:38 pm »
Quote from: maginnovision on Yesterday at 02:12:21 pm
It´s here:
https://www.eevblog.com/forum/testgear/rs-rtm3004-the-swiss-knife-oscilloscope/50/?action=post;quote=1408345;last_msg=2412399
The following users thanked this post: egonotto, maginnovision
Online egonotto
« Reply #6 on: July 19, 2019, 06:24:57 pm »
Hello,
now i found another difference.
The probes in the RTM3k are R&S RT-ZP05 but the probes in RTA3k are R&S RT-ZP10
The probes R&S RT-ZP10 come without BNC Adapter. Why?
Best regards