yeah it sounds similar to the symptoms of "double peeking". I hope it's not because of an old tube
It is sometimes difficult to distinguish double peaking from blooming.
Another possibility is that the CRT has become gassy. If this is the case, just operating it for hours to days with the intensity turned down but not off may restore operation.
To David: Good point. I test it with a 4v 1kHz sinusoid using 1ms/div and 1v/div. The measurement of the time looked correct to me. I didn't pay attention to the value of the peak-to-peak voltage, though. Let me verify it tonight. Do you mean if the vertical axis is way off, then there is still hope for the tube?
If the cathode voltage is low which will cause a dim CRT, then the horizontal and vertical calibration will be off as well. Low voltage will increase the sensitivity; if the cathode voltage is one half of what it should be, then the deflection sensitivity will be doubled.
So if the horizontal and vertical deflection is much higher than they are suppose to be, it indicates that the cathode voltage is low (closer to ground) which will also result in a dimmer CRT.
A more likely problem is that the repetition rate of the test signal is low. Your 1 kHz test signal should be bright at 50uS/div and slower but progressively dimmer at faster sweep speeds. The x10 horizontal magnification function will also dim the trace to 1/10th of its normal brightness.
Also I will measure the voltage when I have a HV probe.
If the horizontal and vertical calibration are correct, then you do not need to bother measuring the high negative cathode voltage because it is correct.