On the catchword “10 ps resolution” we can use the automatic measurements to demonstrate how even entry-level oscilloscopes can do fairly accurate time measurements down to the picoseconds, despite their moderate sample rate.
The test setup includes a stable 500 MHz sinewave, fed into channels 2 and 4 of an SDS2504X Plus (early SDS2304X Plus with 500 MHz option) via a resistive power splitter. First, we use two identical cables, each 200 mm long:

SDS2354X_Plus_skew_200+200mm
The (gate) cursors are there only because I had to use a time gate for the T@M measurements.
Even at 500 ps/div we can barely see the skew of <8 ps between the two signals. This means the total length difference of the two signal paths is less than 1.6 mm.
The distance of the first rising ede to the trigger point at 20% is -2.62 ps on average in the triggered channel and +6.17 ps in the other channel. The difference (skew) is ~7.79 ps on average.
How can we know the resolution? Well, the standard deviation of the skew measurement is 3.38 ps and the peak-to-peak variation is 22.5 ps. The mean value of the measurement statistic is stable.
Another hint is the fact, that the Channel Deskew in the channel menu can be set with 10 ps resolution. Finally, we’ve got the math function interpolate() which allows a maximum of 20. This corresponds to a sample interval of 25 ps and we have no reason to believe that this very same function would not be used internally as well.
Now that we’ve seen a channel skew close to zero, we want some defined delay. Unfortunately, I don’t have the exact same microwave cable in a slightly different length, but a similar one, which is also about 50 mm shorter. I can only assume it’s dielectric has the same velocity factor. Anyway, I’ve replaced the cable to channel 2 with the shorter one – see what we get:

SDS2354X_Plus_skew_150+200mm
50 mm difference at a velocity factor of 0.66 should result in a channel skew of 250 ps. Lo and behold, the measurement says 243,5 ps – 6.55 ps too low, that is not too far off!
If we calculate the length difference corresponding to 6.55 ps at a velocity factor of 0.66, then we get ~1.3 mm. And this seems totally correct. The 200 mm cables have a right-angle plug on one end and a straight one at the other end. The 150 mm cable has two right-angle plugs. I think this explains the 1.3 mm length difference, i.e. the 150 mm cable is a tiny bit longer than initially calculated.
This shows that a serious DSO can do very precise high resolution time measurements, even if it’s only entry level class. Even the SDS824X HD can do pretty much the same in this regard. An SDS7000A on the other hand does a ten times better job and the standard deviation of skew measurements is just ~285 femtoseconds.