Author Topic: DIY 34401A calibration  (Read 18137 times)
Offline baoshiTopic starter
  • Country: sg
    • Digital Me
« on: January 27, 2015, 04:33:38 pm »
Does anyone know, how is the internal wiring of Agilent's 34172A/B short block ?
I have made this but many tell me it is not the correct way for 4 wire short.
Thanks
Sent from my iPhone using Tapatalk
Offline Zucca
  • EE meid in Itali
Re: DIY 34401A calibration
« Reply #1 on: January 27, 2015, 04:53:56 pm »
Why? It should be fine, I would have done it differently anyway. Sense wires carry low current, no need of big copper there.
Offline jlmoon
  • If you fail the first time, keep trying!
« Reply #2 on: January 27, 2015, 05:13:05 pm »
That would be an industrial strength shorting block. indeed.
Recharged Volt-Nut
Offline jpb
« Reply #3 on: January 27, 2015, 05:25:14 pm »
For my Keithley I use two separate standard two pin shorts (see attached photo) which I connect by thick copper wire, it seems to work ok. Doing it with 4 pins as you've done should be fine - the only thing is that you need to be more accurate on placing to avoid perhaps mechanically putting strain on the sockets and I wouldn't trust my own metal working skills to do this - but you look to have done a good job.
NB I put the shorts across force and across sense so the copper wire shouldn't carry any significant current so any contact resistance effects should be negligible.
« Last Edit: January 12, 2019, 12:30:37 pm by jpb »
Offline macboy
« Reply #4 on: January 27, 2015, 06:12:15 pm »
It isn't contact resistance that you need to avoid, it is thermoelectric potentials. With all those plated metal surfaces, that is a lot of possible thermocouples. With the right (wrong?) combination of metals, even a fraction of a degree difference can result in microvolts of offset. This is an issue when calibrating null on a 6.5+ digit meter.
Keithley's shorting block is built on a PCB like attached.
I think the solid copper block is a good idea. I would prefer a plastic standoff instead of a heat-conducting copper screw.
Offline jpb
« Reply #5 on: January 27, 2015, 07:00:21 pm »
I'm no expert, but I'd be surprised if it was a problem for 6.5 digits. I used the approach for the self test on the Keithley which it passed happily so presumably any effect was within the allowed spec.
Out of interest I just experimented a bit with different combinations of connecting the short both directly and via a copper wire between two shorts and with every combination my meter was reading somewhere between 3.3 and 3.8 microVolts (Edit : this drops to around 0.4 microVolts on warming up) which I think is just noise rather than thermo-couple effects.
I accept that you have to be rather more careful with 7.5 and 8.5 digit meters.
« Last Edit: January 27, 2015, 08:47:12 pm by jpb »
Offline quarks
« Reply #6 on: January 27, 2015, 07:07:57 pm »
« Last Edit: January 27, 2015, 07:09:42 pm by quarks »
Offline Dr. Frank
« Reply #7 on: January 27, 2015, 10:50:08 pm »
baoshi,
your short is just perfect!
The big copper plates takes care for low resistance < 1mOhm between hi and lo, and it also takes care for minimum temperature differences between all four plugs, due to its low thermal resistance. This mitigates possible thermal voltages. Also, each plug only has one relevant contact point to the base plate - underneath.
If you wait several minutes after application, you will measure no thermal offset at all.
Look there, that simply bent copper wire is how the 3458A should be shorted, as it's officially drawn in the calibration manual.
That is good for <10nV offset calibration for 8 1/2 digits DCV, and for 7 1/2 digits low ohm (<1mOhm) so that mainly the internal front/rear switch and internal PCB connections can be zeroed during calibration (the 34401A is very similar concerning the front end).
For the 34401A, I soldered that clumsy short on the right, accordingly.
But your design is much better, because it is more simple, and elegantly designed.
Frank
« Last Edit: January 27, 2015, 10:57:52 pm by Dr. Frank »
The following users thanked this post: Synthtech
Offline LaurentR
« Reply #8 on: January 28, 2015, 12:56:21 am »
http://us.flukecal.com/products/accessories/test-leads-probes-and-clips/884x-short
$30 - a tad bit cheaper than the $200 Agilent shorting block    >:D
Or stack three Pomona 2035 (or two 2035 and one or two 3030) - ~$7 each
http://www.pomonaelectronics.com/pdf/d2035_100.pdf
http://www.pomonaelectronics.com/pdf/d3030_1_01.pdf
Offline Vgkid
« Reply #9 on: January 28, 2015, 01:07:54 am »
Dr. Frank, who made those bananna plugs?
  • Country: sg
    • Digital Me
« Reply #10 on: January 28, 2015, 01:26:36 am »
  • Country: sg
    • Digital Me
« Reply #11 on: January 28, 2015, 01:27:17 am »
Thanks everyone.
My idea is the link between the "force" terminals (left) and "sense" terminals (right) need not see current flow. So the connection should be thin. Something like below. Apparently the 3458A calibration manual suggests the same layout.
  • Country: sg
    • Digital Me
« Reply #12 on: January 28, 2015, 03:36:17 am »
Are you writing software to perform the calibration?  If so, any plans to open source it?
Afraid not. See no point to automate it. Just picked up a 34401a and some 9-output reference categorized by a 8.5. Trying to calibrate it.
Offline Dr. Frank
« Reply #13 on: January 28, 2015, 07:47:37 am »
That's stuff from mc, multi-contact.
afaik, it's a copper beryllium bronze, gold, hard plated. LS4 + KT4-B
Frank
Offline HighVoltage
« Reply #14 on: January 28, 2015, 09:42:17 am »
Your copper plate seems perfect.
I wanted to do the same, but never got to it.
The 4mm studs are made by MC (Multi Contact) and I got them already.
So far I have used MC gold plated banana plugs and soldered heavy copper wires on to them, similar to the version by Dr.Frank
It works perfectly.
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #15 on: January 28, 2015, 11:31:54 am »
Not 100% relevant, but since we talking about shorts, I have some PCBs with various shorting configurations.
Fluke have note advising against big thermal mass shorts
« Last Edit: January 28, 2015, 01:23:15 pm by TiN »
Offline HighVoltage
« Reply #16 on: January 28, 2015, 12:36:25 pm »
Hello TiN
Your link to the Fluke document is not working
Could you please fix that.
  • Country: sg
    • Digital Me
« Reply #17 on: January 28, 2015, 05:09:43 pm »
Offline radioFlash
« Reply #18 on: April 02, 2015, 09:30:14 pm »
I got the Fluke 884x-short. Attached are photos which show the layout. There are thick traces between the Hi-Lo sense terminals and the Hi-Lo force terminals (about 60 mils) and a thin trace (about 10 mils) connecting the force and sense sides.
* PCB is .1 inch (2.54 mm) thick
* Weight: 14.7 g
Offline acbern
« Reply #19 on: April 02, 2015, 10:59:04 pm »
I am using these (fluke 884x, keithley has the same under its own name) as well and moved away from thick copper plate shorts as shown in this thread. mostly driven by the fluke app note.
while it is ok (convenient) to use a bare copper wire to calibrate a 3458a, for day to day work with it, the fluke is just better.
for a 34401 the issue is neglectable though, because of its resolution/accuracy.
Offline Tj138waterboy
« Reply #20 on: July 30, 2019, 08:36:57 pm »
Did you happen to make modules for different manufacturer spacings and maybe still have the gerbers?
Online Kleinstein
« Reply #21 on: July 30, 2019, 08:48:54 pm »
The pin configuration is pretty much standard on modern meters. So the Fluke short should also fit Rigol and Keysight.
there is nothing really magic about the simple 4 wire short to adjust 4 wire ohms. It can be carved from a piece of copper clad board (not just leave all the copper), or made with a piece of isolator and than using 3 pieces of wire in an H shape.
The tricky version it the symmetrical 4 contact short that could be turned in any direction. These do exist, but are more tricky and less accurate.
The following users thanked this post: Tj138waterboy
« Reply #22 on: July 30, 2019, 09:07:26 pm »
 :-//Just had a bit of an idea for cal club rd3 with making a standard resistor plug n play set with this design. Would definately need a shield or cover though. Seen this and they aren't same dimensions
« Last Edit: July 30, 2019, 09:16:41 pm by Tj138waterboy »
Offline The Soulman
  • The sky is the limit!
« Reply #23 on: July 31, 2019, 11:48:39 am »
Some more info is here:
https://www.eevblog.com/forum/oshw/4-wire-short/
The spacing between force and sense terminals isn't standardized.
I'm liking your idea tho.
Offline iMo
  • Country: bt
« Reply #24 on: October 26, 2019, 12:11:27 pm »
How the "adjusting" of the 34401A actually works?
Provided I've got 10V source and 10k resistor, and I want to "DIY adjust" my 34401A's 10V and 10k ranges.
I cannot find the details on the procedure, though. I do not want to start to play with the "Calibration" Menu items before having a good understanding what the meter would request me..  :phew:
My current understanding is it will ask me to connect 10V (or 10k) and then to enter (via the front panel keypad) the "exact values", like "10.001.345V" or 9.999.887ohm".
Is my above expectation somehow correct?
« Last Edit: October 26, 2019, 12:14:10 pm by imo »
Offline Dr. Frank
« Reply #25 on: October 26, 2019, 12:20:56 pm »
Please get the service manual, 34401-90013.pdf
The calibration procedure is explained in great detail from page 76 onwards.
Remark: It is necessary to perform the zero calibration first!
Frank
Offline iMo
  • Country: bt
« Reply #26 on: October 26, 2019, 12:32:50 pm »
I've seen that, not sure what it asks for, ie. in the point 5
5. Use ∨ to move down to the “parameters” level and then set the calibration value for the present input value.
Do you dial in the "your number", or you just select "10V" from a list?
Offline Dr. Frank
« Reply #27 on: October 26, 2019, 12:37:31 pm »
that works exactly as the manual number entry of e.g. the Min/MAX, Null Value, etc., i.e. the default is 10.00000, then you select the digit by <> and then increase/decrease the value until it matches your references value.
Frank
« Last Edit: October 26, 2019, 12:39:52 pm by Dr. Frank »
The following users thanked this post: iMo
Offline iMo
  • Country: bt
« Reply #28 on: October 26, 2019, 12:39:47 pm »
And what could be the "spread" of the Vref, Rref you may dial in?
Offline Dr. Frank
« Reply #29 on: October 26, 2019, 12:42:26 pm »
Sorry, RTFM!  :palm:
0.9 ... 1.1 Full Scale, where Full Scale is e.g. 10.00000 V
Offline HighVoltage
« Reply #30 on: May 01, 2022, 10:04:11 am »
After many years of looking out for the Agilent 34172A Calibration Short, finally one showed up and I was lucky to get it. This one came brand new in the original bag.
However, after opening it up, I was a little disappointed.
I did not expect a soldered wire inside.
But it seems this was good enough for Agilent at the time.
Well, at least now we know how it looks inside.
Offline HighVoltage
« Reply #31 on: May 01, 2022, 10:45:43 am »
I just hooked the 34172A short up to a warmed up 34410A in:
- 4W Ohm Mode
- 100 NPLC,
- Offset Comp. ON
- 100 Ohm range (lowest)
It started with 000.000,7 Ohm and took about 1 min to move to 000.000,0 and just sits there very stable.
Offline DC1MC
« Reply #32 on: May 01, 2022, 10:58:50 am »
How much for this gem of an Agilent technology   :-DD  ?
Cheers,
DC1MC
Offline HighVoltage
« Reply #33 on: May 01, 2022, 11:26:40 am »
30 Euro including shipping for.
Well, new price was maybe 10 fold?
At least now we know how the Agilent metrology magic looks inside.
Online tggzzz
« Reply #34 on: May 01, 2022, 11:59:45 am »
- 4W Ohm Mode
- 100 NPLC,
- Offset Comp. ON
With OComp on, I would be worried if it didn't!
What is it like on DCVolts, Null off, Autozero on, 100NLPC ? With Stats on, you could see the min/max (and avg/std dev).
Yes, I like the 34410A's stats display; it often saves uploading to a computer for post-processing.
Offline HighVoltage
« Reply #35 on: May 01, 2022, 02:15:11 pm »
Same display value with Offset Comp OFF
000.000,0 to 000.000,1
About 79 uOhm on average stats display.
Offline HighVoltage
« Reply #36 on: May 01, 2022, 02:23:33 pm »
Here are some pictures of the original FLUKE short block for comparison.
Offline HighVoltage
« Reply #37 on: May 01, 2022, 02:31:42 pm »
Yes, the stats display of the 34410A and 34411A are just perfect.
Voltage readings withe the AGILENT short installed:
- Average: 962 nV
- Max: 1,13 uV
- Min: 743 nV
- StdDev.: 85 nV
« Last Edit: May 01, 2022, 03:03:30 pm by HighVoltage »
Offline HighVoltage
« Reply #38 on: May 01, 2022, 03:03:10 pm »
Same Voltage readings withe the FLUKE short installed on the 34410A:
- Average: 668 nV
- Max: 850 nV
- Min: 508 nV
- StdDev.: 65 nV
Online tggzzz
« Reply #39 on: May 01, 2022, 04:06:52 pm »
My Agilent 34410A with the Bulgarian short was initially -1.14µV to +1.2µV.
After a few hours to stabilise, it was +0.56µV to +1.11µV, mean .85µV, std dev .08µV.
So, pretty similar to yours.