voltsandjolts:
Firmware updated to include relay cycle counters to demonstrate the DAQ commands.
They are just in RAM, so start at 0 on every power-up.
See first post to download.

The cycle counters are accessed from python like this:

--- Code: ---import pyvisa as visa
import time,sys

#py -m pip install -U pyvisa
#py -m pip install -U pyvisa-py     (for the pure python backend)
#py -m pip install -U pyserial

#https://pyvisa.readthedocs.io/en/latest/index.html

#How do I know if PyVISA-py is properly installed? Get info about PyVISA, the installed backends and their options:
# python -m visa info

#On Windows, you may have to uninstall the USBTMC specific driver installed by Windows and re-install a generic driver.
#https://pyvisa.readthedocs.io/en/latest/introduction/configuring.html
#rm = visa.ResourceManager('@py')  # Use pure python backend
#rm = visa.ResourceManager('@ivi') # Use IVI-VISA backend
rm = visa.ResourceManager() # default backend

#print('Available devices: ', rm.list_resources()) # Will show async serial devices (comports) as e.g. ASRL16::INSTR for COM16

print('\nOpening device:')
inst = rm.open_resource('USB0::0x0957::0x2007::MY57007585::INSTR') #For the 34972A on USB
#inst = rm.open_resource("ASRL3::INSTR", baud_rate = 115200) #For the 34970A on RS232 (need to edit port number appropriately 'ASRL?' and check instrument baud setting)

#reading and writing values:
#https://pyvisa.readthedocs.io/en/latest/introduction/rvalues.html

#inst.write('*rst; *cls') #reset to power on state, clear status
# After *rst we now have these settings:
# 1PLC,  5½ digits resolution
# scaling is OFF and scale values are cleared
# sets the scan interval to immediate (0 seconds) and the scan count to 1 sweep.

DEFAULT_TIMEOUT = 5000 # default timeout is 5 seconds
inst.timeout = DEFAULT_TIMEOUT

#print(inst.query('*IDN?'))
#print('Slot 1 card type:')
#print(inst.query(':SYST:CTYPE? 100'))
#print('Slot 2 card type:')
#print(inst.query(':SYST:CTYPE? 200'))
#print('Slot 3 card type:')
#print(inst.query(':SYST:CTYPE? 300'))

print('DAQ Internal relay cycle count:')
print(inst.query('DIAG:DMM:CYCLES?'))

print('Scan card relay cycle count:')
print(inst.query('DIAG:RELAY:CYCLES? (@301:322)'))

#Clear DAQ Internal or the scan card cycle counts like this:
#print(inst.query('CAL:SEC:STAT?'))
#inst.write('CAL:SEC:STAT OFF,HP034970') #unsecure the instrument '970A
#inst.write('CAL:SEC:STAT OFF,AT034972') #unsecure the instrument '972A
#print(inst.query('CAL:SEC:STAT?'))
#inst.write('DIAG:REL:CYCL:CLE (@320)') #clears the relay cycle count on channel 20 in card slot 300
#inst.write('DIAG:REL:CYCL:CLE (@118)') #clears the relay cycle count on channel 18 in card slot 100
#inst.write('DIAGnostic:DMM:CYCLes:CLEar 1') #clear internal DMM relay cycle count K102
#inst.write('DIAGnostic:DMM:CYCLes:CLEar 2') #clear internal DMM relay cycle count K103
#inst.write('DIAGnostic:DMM:CYCLes:CLEar 3') #clear internal DMM relay cycle count K104
#inst.write('CAL:SEC:STAT ON,HP034970') #secure the instrument '970A
#inst.write('CAL:SEC:STAT ON,AT034972') #secure the instrument '972A

--- End code ---