Author Topic: TDS7054 TDS7104 Signal Path Compensation Fix  (Read 2350 times)

0 Members and 1 Guest are viewing this topic.

Offline jkohoutTopic starter

  • Newbie
  • Posts: 2
  • Country: us
TDS7054 TDS7104 Signal Path Compensation Fix
« on: November 03, 2017, 01:49:37 pm »
I have two TDS7000 series scopes that fail Signal Path Compensation.   Everything else works fine, but it cannot pass cal because of the the failure.

I have read all the info about the NVRAM, but it does not look like the issue.  It power up and gets the correct serial number and option codes.

I also looked at the data acquisition board, and nothing looks bad or damaged.

Anyone have any tips on how to fix besides a whole new data acquisition board?

Would a bad NVRAM battery possibly cause the SPC failure? 

Both work working good and passed cal last year.  Only been sitting in the lab and being used, so no physical or storage damage is suspected.

Thanks in advance.

« Last Edit: November 03, 2017, 02:09:20 pm by jkohout »
 

Offline plazma

  • Frequent Contributor
  • **
  • Posts: 474
  • Country: fi
    • Homepage
Re: TDS7054 TDS7104 Signal Path Compensation Fix
« Reply #1 on: February 24, 2018, 10:11:01 pm »
Did you get it fixed?

I have a TDS7104 which usually fails SPC. But after many repeats it passes.
Is it possible to see which channel fails? Maybe components could be swapped between channels to see what is faulty.
 

Offline jkohoutTopic starter

  • Newbie
  • Posts: 2
  • Country: us
Re: TDS7054 TDS7104 Signal Path Compensation Fix
« Reply #2 on: October 31, 2018, 09:13:44 pm »
No, not really.  I reset the NVRAM, and it seemed to help.  Sometimes it passes SPC and sometimes it fails.  If you run it two three times, it eventually passes.  Would be good to figure out, but maybe time to upgrade.

 

Offline Johnny10

  • Frequent Contributor
  • **
  • Posts: 900
  • Country: us
Re: TDS7054 TDS7104 Signal Path Compensation Fix
« Reply #3 on: October 31, 2018, 10:03:14 pm »
I had this problem on a Tek 784D Scope... after changing the acquisition board the error went away.

Reading on this topic I found a Tek Forum customer that found problem in input channel attenuator hybrid.
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 