This is my brief look at the MTR105 Rotating Machine Tester, although from Megger, this is from their Baker Instruments Division that have a long history of manufacturing motor testing devices.
https://megger.com/rotating-machine-tester-mtr105This is a bit more upmarket than the standard insulation testers available, the closest to it is probably the Gossen Metrawatt MetraHit Coil that offers an inter-turn test function alongside an insulation tester and multimeter functions.
The MTR105 offers Insulation testing up to 1000V and 200 GigaOhm resistance level, built into the insulation testing function is DAR, PI, Temperature Compensated and 3 Phase insulation testing. DC and AC voltage measurement up to 1000V with a true RMS function for AC and frequency up to 400Hz. AS part of the voltage measurement it offers 3 phase voltage measurement including phase rotation indication of the supply. To go with this a separate function offers 3 phase rotation indication of a motor, so that it can be matched to the supply.
A continuity function with an adjustable threshold and 2 wire bi-directional resistance test up to 1 MegOhm with a 200mA test current for resistance values below 4 Ohms. It also has a 4 wire resistance mode included for testing larger motor windings from 1 milliOhm to 10 Ohms with a 0.01 milliOhm resolution. An LCR meter is included using either 120Hz or 1kHz test frequencies, capacitance from 0.1nF to 1.0mF and inductance from 2mH to 1H. It has a temperature function with a T-Type thermocouple that works in conjunction with the temperature compensated insulation test mode.
Test data is saved on board the instrument to memory slots that you have to provide a name for, al individual tests are then saved under the slot selected. There is no direct connection to a computer, the data is downloaded to a USB stick as csv files for each individual test. This can be read by the Megger PowerDB Software and pulled into a certificate or pulled into a spreadsheet program such as Excel to create your own certificates.
Couple of pictures of what arrives with the meter, set of 3-phase leads, 3 test probes, 3 grippers, t-type thermocouple, 4 wire kelvin leads with crocodile clips a hanging strap, all to go in a soft case. A USB stick to download data to, calibration certificate and quick-start guide are also included but not shown in the pictures. A charger with a set of 6 NiMh AA cells and 4-wire duplex pin leads are available as extras that I have purchased alongside the meter.
I have been asked for my test data from it, so I will also add that in as a pdf.
Next post will show some picture of the instrument tests.