Author Topic: DMM6500 or 34465A  (Read 8642 times)
Offline TrickyNekroTopic starter
  • Country: gr
  • Drinking Frappé on the way to Isomnia!
    • Youtube channel, electronics and more ;-)
DMM6500 or 34465A
« on: June 18, 2021, 06:10:07 pm »
Hi!
I am thinking of going the extra mile and buying a proper bench DMM. I am already going through older forum posts but I would like to ask this question again now,
2 years after the last post I saw on the forum from googling around.
I do not really need the low Ohm range of the DMM6500 for what I do. Data logging is of greater interest.
I lean towards the 34465A due to availability in my country, better financing and support in my country.
Performance wise I lean towards the DMM6500 due to higher sampling rates but I do not know how well you
keep both precision and high sampling rates, it´s more of the impressions the DMM6500 left me from watching
reviews rather than concrete data, that´s why I also post here, to listen to experience.
I will not shy away from a third option that provides better performance in this price range, not only performance wise though
but also UI wise, I really want to be able to do some nice scripting and datalogging over ethernet.
What´s your opinion? I have not found a Youtube video yet that breaks down all the options pros and cons, would be really
happy to watch some!
Use is going to be a mix bag of hobbyist and professional use. Now I will be using it as a hobbyist / student, but I am looking
forward to use it one day professionally, target is low voltage devices, 48V tops, anything from light industrial actuators to IoT.
Thanks again for your replies in advance!
Best Regards,
Lefteris
If you are an engineer and you are not tired...
You are doing it wrong!
Online nctnico
Re: DMM6500 or 34465A
« Reply #1 on: June 18, 2021, 06:19:02 pm »
My general impression is that the DMM6500 has more features and could be considered better in many ways. OTOH it seems that the DMM6500 has some firmware issues but I don't know whether these have been ironed out yet. I own a Keysight 34461A myself but probably would have liked the DMM6500 better if it was available when I bought the 34461A.
The following users thanked this post: TrickyNekro
Offline Kleinstein
« Reply #2 on: June 18, 2021, 07:22:06 pm »
The higher speed mode of the DMM6500 use a different ADC with lower accuracy. The Bandwidth is limited by the input amplifier, so that the full sampling rate is of limited use. I don't know the speed of the input amplifier in the 34465, but is may be similar speed.
The 34465 is quite a bit lower noise, when it comes to just the ADC - at least at lower speed.
Normally the 34465 is quite a bit higher cost. Cost wise the DMM6500 is more comparable to the slower 34461.
The DMM6500 has the option to use a scanner card.
Online daqq
  • Country: sk
    • My site
« Reply #3 on: June 18, 2021, 08:01:15 pm »
I have a DMM6500 at work, best multimeter buy ever. Great, responsive and intuitive user interface and what's also great is the option for expansion cards ( https://assets.testequity.com/te1/Documents/pdf/keithley/2000-SCAN-manual.pdf ) - this turned out to be VERY useful for us.
In terms of parameters it's OK for me.
Not saying that the 34465A is a bad instrument - I've worked with the 34461A, which at a quick glance is very similar to the 34465A. Good device, but I'd go for the DMM6500 any day.
The following users thanked this post: TrickyNekro, samofab, tooki
  • Country: gr
« Reply #4 on: June 18, 2021, 09:18:32 pm »
Suspected as much, I will have to go through the specs and multiple reviews one by one.
True, the difference is about 500€ which of course is something to consider as it´s almost 1/3 of the price of the 34465A.
But over 6 months payments interest - free, these extra 85€ are not that much of a pain.
Giving though 1100 - 1200€ just like that for the DMM6500 is going to be very painful. Plus I have to be buying from Germany probably.
There is supposedly a local distributor but their site left me with the impression that it was left a bit behind. Sure that´s only my impression
but you know... gut feelings...
But the ADC being less noisy is actually reason enough to prefer one over the other. I need to find the use cases where on is better than the
other really to be able to make a decision.
I think that´s the DAQ6510 ( or is the DAQ6510 just a DMM6500 with this card? ). Which then goes about the same price as 34465A so meh pretty much the same.
Anyways, thanks for your reply!
« Last Edit: June 18, 2021, 09:22:08 pm by TrickyNekro »
  • Country: gr
« Reply #5 on: June 18, 2021, 09:24:58 pm »
Is something specific that makes you say that? This feature, that feature etc., I need to dig in a bit.
Thanks
Online daqq
  • Country: sk
    • My site
« Reply #6 on: June 18, 2021, 09:41:09 pm »
Thanks
General feel, how the graphing is done, user interface responsivity - somehow they made me like a touch screen on a measurement instrument, a thing unheard of! The numeric entry is usable... lots of UI stuff. But also the fact that it's damn fast and damn responsive.
And the scanner card option is really a great feature. Tad pricey, or you can build your own:
https://www.eevblog.com/forum/circuit-studio/example-project-relay-scan-card-for-k2000-dmm/
Plus the parameters are good.
Nope. The DMM6500 supports Keithley scanner cards (see 2000-SCAN and others).
Online Someone
    • send complaints here
« Reply #7 on: June 19, 2021, 04:58:09 am »
The accuracy and precision at different sampling rates are relatively well documented in the data sheets, I think it gets more complex as the DMM6500 has a different "mode" for the fastest rates.
Don't forget that Keysight have models up and down from the 34465A, so there are many points to compare if you are pressed on budget. For example I have used the very similar 34461/65/70 meters and never found any killer app for the higher sampling rates, as high speed current and voltage could be handled by suitable scope probes. Equally for low current use I skip right past multimeters and use an SMU to eliminate burden. So whats best value is really in context of what other equipment you have, and the very specific measurement needs of your applications.
It took some time and many false starts, but the best touchscreen interfaces are now ahead of button/knob only. As you say, its the speed of the interactions that ends up critical for UIs, some button/knob interfaces fail there too!
  • Country: gr
« Reply #8 on: June 19, 2021, 09:02:15 am »
Exactly because I have a scope, I am leaning towards the 34465A. Right now, I find the 34461A and the 34465A with a ~500€ difference, because though I can pay over 6 months, it doesn´t really dent me.
Or else I have to collect the money over the next 6 months to buy the DMM6500 and that´s at least 6 months without the instrument which could be a thing as then I would have to run between many places
to have access to measurement level DMM, plus the temptation of drinking that extra coffee or beer.  :popcorn:
The DMM6500 from reviews so far has me a bit more "wow"-ed. But I do not want to buy a 1k+ class instrument on "wow" factor.  :-//
Still I wish to go the extra mile cause I know this will become part of my profession, so that´s only early investment, a worth one though.
I will go through old Dave videos and return with more specific questions. I really can´t remember if he made a comparison of the two. Sure
he has reviewed it, he and / or the "Signal path" guy ( name difficult to remember sorry! ).
But on the Signal path channel, I can not remember a direct comparison of the two.
Ah well...
Online HighVoltage
« Reply #9 on: June 19, 2021, 09:14:58 am »
I have both and I like them both and really could not say that one is more recommended than the other.
I like the many direct physical knobs of  34465A but I also like the very intuitive touch screen of the DMM6500
It also depends on what you want to do with the DMM.
The 34465A is my daily "go to" instrument on the bench for most work.
It is very reliable and at least on my unit the calibration is rock solid even after a few years.
In order for you to decide, you might get a loaner unit of each and test them for a week or so for free.
  • Country: gr
« Reply #10 on: June 20, 2021, 12:09:29 am »
Looking at this old Dave video, what I understood, at least from the video, is that the 34465A is a 34470A waiting for a LTZ reference to happen.
So if I would go to the Keysight route getting the 34465A over the 34461A is worth it.
Video here:
  • Country: gr
« Reply #11 on: June 20, 2021, 12:57:46 am »
I see, I see. Found out the details about it also. To be honest though, I play around labview and know how to program a microcontroller or two. I could build such a card myself externally,
regardless of the DMM controlling it with even a simple Arduino or any other USB enabled microcontroller. That being said, I do not really need the multiple inputs from such a scan-card.
I would totally understand the need in a production environment with automated testing, where you do not want to play around.
Of course on the other hand, both seem to have nice labview drivers so given the need automation is not going to be an issue.
I will go through Dave´s DMM7510 review now. Try to find how hard the differences between the 6.5 and the 7.5 are.
Cheers!!!
« Last Edit: June 20, 2021, 01:05:07 am by TrickyNekro »
  • Country: gr
« Reply #12 on: June 20, 2021, 01:58:43 am »
Well I am going through the specs and it really makes it a hard choice.
The 34465A seems to be more accurate apart from the low Ohm range, which in all honesty I do not care all that much, I do not build shunts every day and even when I use them
there are other ways to calibrate them.
Generally with their "truevolt" trademark, they got better DC and AC stability. And a big difference also comes from the AutoCal feature, especially for the long term stability.
The DMM6500 has the interface, "wow-factor" low Ohm, higher diode testing voltage and so on going for it. If I was comparing with the 34461A then the DMM6500 would clearly
be better, much better as a feature and performance ratio.
Of course the 34465A costs quite bit more but that ACAL... that ACAL really makes a difference long term. Do I really need it? Yeah, I guess kinda, yes.
The DMM6500 has also better programmability and scripting, but it would be yet another scripting language I got to learn and then I will forget.
Since both companies support labview then I would much prefer to set up a labview program given the need for automation. Also from some random comments
I saw that people tended to be more happy with BenchVue and it is provided for free from what I understood. A confirmation on this would be great.
One great thing that the 34465A has, is the low current range. Since I work with microcontrollers and have also done some IoT before, current draw datalogging seems really
possible.
It is a really hard choice really. Between the DMM6500 and the 34461A, I would have gone with the DMM6500 hands down. Now it´s much more tempting to go with
the 34465A after I have done the comparison, but it´s also tempting to keep the money difference and equip my lab with a function gen, or something similar that is still
missing.
Tough call really, if the dealership comes through with a good enough offer, I will go for the 34465A. Else most probably I will go with the DMM6500.
If there is something I absolutely have to know, speak now or else be silent forever!  :P
Cheers!
Lefteris
Offline bdunham7
« Reply #13 on: June 20, 2021, 05:13:12 am »
I think  you might want to look at those specs again, perhaps post what you are looking at.  I don't see the 34465A as being more accurate.
One issue I have run into (with a different meter) is that while the DMM6500 uses a different current shunt for each of the lower ranges, the 34465A appears to use the same shunt for all 3 of the lowest ranges and recycles some others.  The DMM6500 has 7 different shunts, which is a big commitment to accuracy.  When you reuse a shunt and apply gain, you get some noise--it's unavoidable.  The 34465A probably has better digital trickery to smooth that noise, but their specs for the 3 lowest current ranges look suspect to me--I don't know why they include the comment "(typ)".  I'd ask.
Also, I find an accurate 10R range to be very useful.
These are just two observations.  I don't have either meter so I can't say much else.
The following users thanked this post: ldmpam
Offline Kleinstein
« Reply #14 on: June 20, 2021, 07:05:58 am »
From the specs of the 34465 it looks like the 10 mA range is the same shunt as 100 mA, just with additional gain. This is visible from higher percentage error ralative to the range. The small ranges look more like they use a TIA instead of classical shunts. With just shunts one could hardly get that level of accuracy with so little burden. For stability a TIA still need some series resistance and thus the burden is not zero, but can be quite small.
The typical remark with the small ranges is likely mainly from leakage of the input protection, restricting it too room temperature.
They explain it at the front. A 80% confidence level would imply about an additional factor 2 in comparisom with the normal 95% or 99% level - though this may not be a normal distribution thing, more like a question of testing / rejecting.
I agree that the 34465 is better in digital handling the data to reduce noise. However this is more like a flaw on the Keithley side, that tends to add some LF noise with integration / averaging over some 0.5 - 50 seconds.
ACAL function can help to get very good long term stability (e.g. > 5 years) and allows then manufacturer to use cheaper parts in some places. It especially helps when comparing reading measured in different ranges (e.g. 1 V and 100 V) as ACAL bings that ratio essentially back to the 24 h specs, while without ACAL it would be more like the 1 year specs, unless directly after calibration with adjustment (or using measurend correction factors).
AFAIK CAL in the 34465 does no include all ranges ( espeically AC).
Online HighVoltage
« Reply #15 on: June 20, 2021, 07:35:34 am »
The ACAL on the 34465A works well but you only need it, if you have large temperature changes in your lab.
I don't think it has any influence on the long term stability.
« Reply #16 on: June 20, 2021, 07:48:56 am »
snippity
Cheers!
Lefteris
If you seriously need low Ohms accuracy then a dedicated MilliOhm meter will run rings around the 34401A, 34461A (the two I have that are kept in Cal) and likely the 65A and the DMM6500 as well. So I would tend to cross this item off you list all together as a consideration in some ways and put the $4-500 saved toward a dedicated Milliohm meter. The Quadtech I scored some time ago is my go to and not my 6 1/2 or 7 1/2 digit meters (Advantest in the fleet too) for below a few Ohms.
Without rechecking the complete spec the 34465A is a hard sell for the extra $ unless you really need the high speed reading side of it the accuracy differences are fairly minimal.
Overall the question should be is how important are the last 10PPM to you now and into the next 5-10 years and if they are keep saving and buy a 3458A (got one of those in the repair que) :-DD
  • Country: gr
« Reply #17 on: June 20, 2021, 10:05:48 am »
That´s the thing. I do not really. I much prefer a good DC current accuracy, to calibrate my shunts with their respective amplifier rather than the exact resistor accuracy. So I get I am a bit sold there.
I will leave it for a day in greener pastures, when my kidneys are not at risk of being sold.
  • Country: gr
« Reply #18 on: June 20, 2021, 10:08:40 am »
I will clarify this, but that´s ok even so. Summers here are brutal and you can´t be hugging the A/C all the time.
Online daqq
  • Country: sk
    • My site
« Reply #19 on: June 20, 2021, 10:40:55 am »
Actually I had to log 4 temperatures from PT1000s of a device for a few hours. It was not part of an automated test jig, just an experiment.
Also, have you considered older meters? If you are concerned about user interface stuff, meters like the venerable 34401A, Keithley 2000 or similar can be found reasonably cheaply. In terms of parameters they aren't bad at all.
  • Country: gr
« Reply #20 on: June 20, 2021, 10:45:02 am »
I was comparing from these two:
https://download.tek.com/datasheet/1KW-61315-0_DMM6500_Datasheet_052119.pdf
https://www.keysight.com/zz/en/assets/7018-03846/data-sheets/5991-1983.pdf
Granted the differences between the two are not earth shattering and maybe not worth the extra 500€.
Yes, that is something to "worry" about. Although that being said, this is the exact reason they have better burden voltages and I had it before that I had a problem with a device browning out / resetting
when exiting sleep mode.
So you can either take it as a bug or as a feature really... If they meet the spec, I wouldn´t really care, in the end of the day, even with the "caveats" these are top of the line companies, we are talking about.
  • Country: gr
« Reply #21 on: June 20, 2021, 10:50:19 am »
I am very sure that the day will come that I will say, "darn I should have bought the other". But that´s probably why one day I will try to buy both :P
Ah, yes. But you know it is not the same. I know it is probably the illusion, and you are 100% right, but let me be happy, ok?  :scared:
Offline bdunham7
« Reply #22 on: June 20, 2021, 01:38:11 pm »
What I'm not seeing is how you come to the conclusion that the Keysight is better at all.  They're quite clearly head-to-head competitors, but the Keithley appears to be equal or better on everything except low and high frequency AC voltage.  I didn't read every spec--they both are probably more than good enough on most ranges--but you may need to pick the ones that matter.  I would be very wary of the asterisked lower current ranges.  Perhaps you were looking at some other specs that matter more to you?  If you look at features, they're very similar, but the Keysight does have the 1G resistance range while the Keithley has 1R and 10R and a better selection of diode ranges.  And of course the scan card option.  ACAL vs autozero is another issue that I think is mostly a non-issue in most cases, but I wouldn't regard ACAL as an advantage unless the meter is used in widely varying (outside the 18-28C window) temperatures.
Unless they have implemented some very special low-noise amplification in the lower current ranges, it's a bug-period.  You aren't going to be measuring individual picoamps because you are limited by the noise, accuracy and input bias current of the voltage section.  With a shunt for each current range, it is easy enough to get lower burden voltages by just selecting the next range up.  With the Keysight setup, you don't have a choice.  As I said I'm sure they are both excellent products, but I would have to ding Keysight here for a bit of marketeering and perhaps specsmanship.  I don't like the "(typ)" denotation because I think they are quoting a non-guaranteed spec.  If I needed thos ranges to be accurate for my work, I'd be pressing them to explain the details of that spec or finding someone that has reviewed and tested the meter in those ranges. I think their blurb on 'low input bias' is a sham with the bar graphs showing how they are better than the competition but they won't quote an input bias spec--and ignore the decades-old standard Fluke 8846A with its typical 15pA and guaranteed <30pA bias current.  Comparing your high-end 6.5 digit DMM to brands 'A, B and C' when none of the three are Fluke is a bit deceptive, IMO.  And if they are comparing their 'typical' measurement to a guaranteed spec, that's deceptive too.
Anyway, none of that should stop you from buying either meter if you like them and I'm not advocating for either.  There are other non-spec related issues as well, but since I don't have the meters I can't comment on the UI or things like that.
« Last Edit: June 20, 2021, 02:19:22 pm by bdunham7 »
  • Country: gr
« Reply #23 on: June 20, 2021, 03:15:55 pm »
I got the feeling that yesterday night I confused the charts between them, or something. They are clearly head to head, in some cases one being slightly better than the other but nothing tragic.
In a day I could have a variation of 5 - 6 °C in the lab, especially in summer but it´s not the be all, end all of the game here.
The diode range is really a thing, but dirty old me says I will just hook my power supply in series in case I need the extra voltage.
I will try to find reviews on the lower current ranges. It is difficult for me to test them. I do not expect to count pA either to tell the truth. I need better accuracy on the 1uA - 100mA range than anything else,
with probably a single shunt that will not switch when a device wakes up.
Forcing a range is something, I have not thought really and that´s actually a good point to gravitate towards the Keithley.
I really think it will come down to the dealership making me a good offer here. 500€+ difference is a big one. I could get me a second Rigol bench power supply or function gen for the same money.
Which is a lot if you think about it. I´ll see the next days really...
Offline Hexley
« Reply #24 on: June 20, 2021, 05:30:53 pm »
... target is low voltage devices, 48V tops, anything from light industrial actuators to IoT.
If you plan to measure power supply current in a typical IoT device, it will be useful to use the 10A range of your 6.5 digit meter in order to get a very low burden voltage. The 34465A spec sheet alludes to this in the DC/AC current burden voltage spec, where they show that the burden voltage of the 1A range can be reduced from "<0.7 V" to "<0.05 V", with a footnote for the latter saying to use the 10A input range.
(The DMM6500A, though it only shows the burden voltage at 1A as "< 0.55 V" for DC and "<0.75 V" for AC, could be used in the same way, of course.)
The 10A range is for me the most-used current range for exactly this reason -- low burden voltage. So it is worth noting a small difference between the two brands -- the Keithley has the 10A input on the rear panel, while the Keysight has it on the front panel.
Since I frequently need to access the 10A terminal, I find its front-panel location in the 34465A quite convenient in this regard.
Just passing along this bit of real-world experience about a difference that might not otherwise leap off the data sheets that you are comparing. :-)
The following users thanked this post: skander36