To help try and resolve this issue, I waited until I was back at work today. There is a suite of Gedore torque testers there and I selected the 0.5 to 5 Nm model. It looks like a bigger and more expensive version of the one linked to below:
https://www.gedore.com/en-de/products/torque-tools/torque,-measuring-testing-appliances/torque-testing-appliances/capture-hub/ch-5---3124053I did try and remember the model number but I've forgotten it. The datasheet said 0.5% accuracy across most of the range. It measures and holds the display at the peak torque achieved at the break point of the torque tool under test.
I tested three Agilent 8710-1765 torque tools taken from 85033E VNA cal kits. I held them the same way as recommended by my N4431B-60006 Ecal module manual as seen on page 4-19 here:
https://www.keysight.com/gb/en/assets/9018-03493/reference-guides/9018-03493.pdf?success=trueThis means holding it between finger and thumb at the tip of the tool.
When I do this, I get really repeatable results each time. There is a slight difference between each torque tool but not much. When tested, they ranged from 0.897Nm to 0.917 Nm but probably none of them have been adjusted in many years.
In terms of repeatability, I could typically achieve better than +/- 0.01 Nm if I really tried to hold and rotate the torque tool in exactly the same way. This was quite impressive. I also saw about a 0.025 Nm difference if I reversed the tool as the hex cutout is at an angle.
If I then tried the same at halfway down the gold handle of the 8710-1765 then the reported torque was at about 1.07 Nm. If I tried at most of the way down the handle (where it says 0.9 Nm) then the torque went up to about 1.3 Nm. However, it's unlikely anyone would use the tool like this.
If I held it the way most people would hold it fairly near the tip using several fingers and thumb (as shown as the 'wrong' way to hold it in that guide doc) then I couldn't get repeatable results. A lot depends on how the fingers or thumb apply the pressure. I could see anything from 1.0 Nm to 0.8 Nm. Usually it showed 0.85 Nm to 0.95 Nm which is quite close to the 0.9 Nm spec for this torque wrench.
I did also notice that if I held the torque tool like a golf club with my thumb along the shaft and several fingers wrapped around it, the breaking torque could sometimes go below 0.75 Nm.
So in summary, if you are a nerd and you really do want to achieve repeatable torque settings that are impressively close to 0.9 Nm then hold it at the tip between finger and thumb as recommended by Keysight.
However, if you are an engineer just doing everyday SOLT calibrations with a sub 8.5 GHz VNA then I don't see a problem holding it the 'wrong' way because the error in the torque isn't that significant in my opinion.