Hi everyone, I'm the owner of DP811A, I installed the v1.11 firmware.
I made the calibration of the V-DAC and the I-DAC output voltages and now all wrong.   :-BROKE I tried for three days to redo the settings, to no avail. Then I read here on the forums that the 11.1 have a bug.
Now I have an old iron box unusable paid 750 Euros. :palm:
Someone can 'help me to solve this problem?
I contacted RIGOL EU Engineer Mr Thomas told me that I should not touch the calibration and now so I have no warranty and I have to send them by paying the repair.  :-- To me this does not seem fair, I did not do the problem, but they did their damage with a firmware bug.  >:D
 :-/O Someone SCPI commands for calibration, or any other solution to seat the power supply?  :-DMM
THANKS!  :)
Offline TooOldForThis
  • H: 42.576MHz/Tesla
Re: DP811A ( DP800 ) SCPI Command Calibration for PC
« Reply #1 on: December 02, 2014, 02:33:36 am »
Hello cidak,
    Use the SCPI calibration instructions posted  Here
They should work on the DP811A too.  But the DP811A has different calibration points.  Use the cal points in the document I attached.
« Reply #2 on: December 02, 2014, 02:45:26 am »
The DP811A can output 10.5Amps.   So unless you have a real brute for a DMM, you'll have to measure the upper current ranges by measuring the voltage across a low impedance power resister attached between the + and - terminals.
« Reply #3 on: December 14, 2014, 03:50:49 pm »
  :clap: Thank you very much for providing information on how to calibrate  :bullshit: the DP811A Power Supplies.
I for one appreciate this very much, and I'm sure everyone else here does also. :-+