Hi All, can I get your $0.02?
I have run into a problem with my use of a BK 8500 E-Load: At low input powers, the load is unable to stabilize.
The experiment works like this: attach power supply to e-load and set to 3V w/10mA limit. Set the e-load to 100 ohms.
Of course, this circuit should require 30mA. When given more than 30mA, the BK is just fine. But, when given less than the right ohms-law amount, the load gets pretty wild.
We also have a Chroma load and it does a good job with this... however, it is considerably more $$$. I was hoping for something a bit smaller (and with USB/RS232/GPIB/etc interface)
My questions are:
1. General thoughts?
2. Although I believe I have ruled this out, is there a supply to load interaction at play here? I am using a Rigol DP832A for the supply.
c. Does anyone have a load that can pass the above test? ... or can recommend a load that can?
4. Anyone have a BK 8600 who wouldn't mind trying this out?
-
Can you try a cap (1uf ceramic // 10/47uf electrolytic) directly across the input sockets on the load ? This has cured instabilities for me in the past before.
-
Can you try a cap (1uf ceramic // 10/47uf electrolytic) directly across the input sockets on the load ? This has cured instabilities for me in the past before.
Thanks Chris. I will give it a shot and let you know
-
I've found some loads can be problematic in constant resistance modes. Constant current is the more stable in some cases.
I doubt it's the power supply not liking a low current. To try this you could connect a high value resistor to the power supply and measure the output voltage.
-
I've found some loads can be problematic in constant resistance modes. Constant current is the more stable in some cases.
I doubt it's the power supply not liking a low current. To try this you could connect a high value resistor to the power supply and measure the output voltage.
This is part of what I did to prove to myself that its not the supply... unfortunately, I am stuck with CR... although I am tinkering with Constant Power
Thanks again